Help, how do I enter calibration mode?
daisizhou:
Hallo every one
I was repairing an analyzer recently.The fault phenomenon is that the pressure deviation exceeds the allowable value.
I read the user manual and found only instructions for oxygen calibration,There is no detailed calibration procedure for pressure sensors or flow sensors.
By reading "User Communication Interface" I found some clues about calibration,The calibration date and calibration ID are marked on it.
In the normal user interface, I do not see the calibration date and calibration ID.
So I got the GUI firmware and read it with "Winhex" software,Found some information about calibration, but it is still incomplete.
Due to language differences, I am not sensitive to English characters.
If you can find some clues, I look forward to your help, thanks
jasonquin:
I think that is a multi-language menu section, Italian and Spanish or German in the picture.
daisizhou:
Yes, there are indeed many languages to choose from, but I hope to find a way or method to enter calibration mode
jasonquin:
Fluke allows you to calibrate oxygen sensor only because it should be calibrated at the beginning of everyday. If something else fails I think Fluke wants you to send the device to Fluke Biomedical Service Center, as the manual says if any other calibration is done by a non-Fluke qualified technician, you lose the warranty.
daisizhou:
I understand that this will result in the loss of warranty, but the warranty has already expired.
I'm usually not used to contacting manufacturers because they are unfriendly and even unwilling to provide service manuals.
Due to the high maintenance cost, I can only choose to have it repaired myself or by a third-party