Ohh this puppy is driving me crazy!!!
When the meter arrived in my hands it showed the following faults: 302.1 302.2 600.1 600.2 601.2 (the latter maybe because I does not short the amp probe).
I decided to change the relay k102 (because? read first post) but after this the problem was not solved. So I put back original relay in place.
At this point I decided to follow what was stated in manual for fault 302.1 but some shit happens. After a while the 4 wire ohm started to read higher.
I run autotest again and now I have also error 100.2 (and many others related to this).
After this "shit" the 4W ohm is reading higher:
- 100 and 1k ohm: +5%
- 1M ohm: +30%
- other range: no consistent reads
Strange is that DC is still OK.
So I decided to go ahead and try to solve the 2wire ohm problem when at a certain point it started to measure similar values to the 4W ohm.
I just had time to run the 300 series tests with no fault and probe some voltage before the miracle to vanish.
During this time I notice that I have expected voltage (~ 0.490V for ~500 ohm resitor) to the input/output of U113. This is almost the same value I have with 4 wire and same resistor now.
Don't know ... any idea?
It's very hard without schematics to figure what's happening.
Nobody has any sketch of analogic section of this meter?