| NY2KW:
I would appreciate help understanding correct use of a high voltage differential probe specification (MicSig MDP702) maximum input differential Voltage versus maximum voltage to ground. I want to use the probe as a single end active probe to trace small signals (~ 3MHz) in a RF tube amplifier with plate voltages ~ 200-300VDC. I believe my problem is in one of the LC tuned circuits between stages and there is full plate voltage at certain points I need to test. My basic question: am I correct that I can the probe at the 20x setting for single input referenced to chassis ground based on the CAT II limits or must I use it at the 200x setting (which has a much higher noise floor)? My signal levels at some point are in the 100mV range. Maximum differential input ratings: @ 20x : 70V (DC+AC PK) @ 200x : 700V (DC+AC PK) Max. Voltage to ground rating: CAT II 450V Thanks in advance Jerry |