Products > Test Equipment

How do I switch to Constant Current instead of Constant Voltage on RD6018?

(1/4) > >>

sofakng:
I'm trying to test an LED with a max current of 100 mA.

The RD6018 shows [CV] at the bottom of the screen so I'm assuming it's in Constant Voltage mode so how do I switch to Constant Current?

I've already destroyed one LED because of my misunderstanding of the bench power supply.  I had V-SET to 5.00v and I-SET to 0.9A (and no OCP).  I'm confused though because I have current set to 0.9A but I'm guessing that only works in Constant Current mode?

Fungus:
I doubt it has constant current mode.

For an LED you can set the voltage much too high then use the current limiter. Set it to 100mA.

wasedadoc:
0.9A is 900mA.  That is 9 times the 100mA max of the LED.

Martin72:
Hi,

A supply is in it´s constant current mode when you have reached the pre-adjusted current.
Until this it´s in constant voltage mode.
For your LED you should set the current to 100mA before.
BTW, what LED is it?

sofakng:
Ahhhh... thanks.  I'm an idiot.  0.9A is 900mA as mentioned by wasedadoc.

The LEDs are the SFH 4547.

What is the expected behavior of the following settings?

V-SET: 5.0v
I-SET: 0.1A (100mA)

The LED has a forward voltage of 1.5V so would the PSU reduce the voltage from 5.0V to 1.5V and go into Constant Current mode to limit the current to 100 mA?

Also -- When I measure the LED with my multimeter (Fluke 87V) using the diode mode, it shows 1.1V.  The specifications show 1.5V so I was expecting to see that?