Hello, I bought a Siglent SDG1032X signal generator and a Siglent SDS1104X-E oscilloscope in 2019, but playing with telnet I deleted the calibration file and the serial number of the SDG1032X, now that I have just bought a Siglent SDM3055 5-digit bench multimeter and a half this week, I want to take the opportunity to calibrate the SDG1032X.
I have written an email to Batronix where I bought it to explain how to calibrate it and/or to provide me with the calibration scripts. They replied that I should consult Siglent directly. I consulted Siglent a week ago and they have not responded.

Can someone here explain to me how to calibrate the SDG1032X and give me the calibration scripts, thanks in advance.

And by the way, you know how to calibrate the SDS1104X-E and if you explain it to me, it would be perfect.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: tautech on October 14, 2023, 06:11:00 pm
Hello, I bought a Siglent SDG1032X signal generator and a Siglent SDS1104X-E oscilloscope in 2019, but playing with telnet I deleted the calibration file and the serial number of the SDG1032X, now that I have just bought a Siglent SDM3055 5-digit bench multimeter and a half this week, I want to take the opportunity to calibrate the SDG1032X.
I have written an email to Batronix where I bought it to explain how to calibrate it and/or to provide me with the calibration scripts. They replied that I should consult Siglent directly. I consulted Siglent a week ago and they have not responded.

Can someone here explain to me how to calibrate the SDG1032X and give me the calibration scripts, thanks in advance.
Maybe our recovery packages can help.
Send PM with your email address.

Quote
And by the way, you know how to calibrate the SDS1104X-E and if you explain it to me, it would be perfect.
Not available.
Update to latest FW and OS then run Self Cal after prescribed warm up period.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 14, 2023, 07:30:47 pm
Thank you very much I await your PM
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 16, 2023, 07:35:04 pm
Hello Tautech, thank you very much for the recovery kit, I put it on a USB and it ran, it seemed to work, but it stayed the same since it already worked and I already had the latest version of the software.
That is to say, it stayed the same as it was. :(
Recovering the old telnet thread on the SDG1032X, I was able to manually modify the file by Telnet and put my real serial number back into it.
But calibrate it I still don't know how to calibrate it...
My reference is the SDM3055 that I bought a week ago and it comes calibrated.
Well, putting the SDG1032X in:
C1: DC 1mV the SDM3055 measures 2.762mV and the UT-71E measures 2.7mV
C2: DC 1mV the SDM3055 measures 2.834mV and the UT-71E measures 2.8mV
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 16, 2023, 07:43:09 pm
If anyone knows how to calibrate it and explains it to me, I would appreciate it.

In the old service manual it refers to a calibration script and gives a link that no longer exists. I think years ago this calibration script was available, today there is no trace of it, and I don't know if the current firmware allows it...
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: tautech on October 16, 2023, 07:56:10 pm
Try restoring factory settings from the Help menu.
P34
https://siglentna.com/wp-content/uploads/dlm_uploads/2020/12/SDG1000X_ServiceManual_SM0201X-E01B.pdf
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: KungFuJosh on October 16, 2023, 08:18:22 pm
My reference is the SDM3055 that I bought a week ago and it comes calibrated.
Well, putting the SDG1032X in:
C1: DC 1mV the SDM3055 measures 2.762mV and the UT-71E measures 2.7mV
C2: DC 1mV the SDM3055 measures 2.834mV and the UT-71E measures 2.8mV

How are you outputting 1mVDC from your AWG? The signals are generally AC, with the option for a DC offset.

Thanks,
Josh
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 16, 2023, 08:20:26 pm
Try restoring factory settings from the Help menu.
P34
https://siglentna.com/wp-content/uploads/dlm_uploads/2020/12/SDG1000X_ServiceManual_SM0201X-E01B.pdf

Thanks Tautech I have done it, but it's still the same
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 16, 2023, 08:23:39 pm
My reference is the SDM3055 that I bought a week ago and it comes calibrated.
Well, putting the SDG1032X in:
C1: DC 1mV the SDM3055 measures 2.762mV and the UT-71E measures 2.7mV
C2: DC 1mV the SDM3055 measures 2.834mV and the UT-71E measures 2.8mV

How are you outputting 1mVDC from your AWG? The signals are generally AC, with the option for a DC offset.

Thanks,
Josh

Being on channel 1 (for example), press the "waveform" key then go to page 2, click on "DC", and with "DC Offset" you choose the voltage.
It leaves you from -10vdc to 10vdc
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: tautech on October 16, 2023, 08:26:34 pm
If anyone knows how to calibrate it and explains it to me, I would appreciate it.

In the old service manual it refers to a calibration script and gives a link that no longer exists. I think years ago this calibration script was available, today there is no trace of it, and I don't know if the current firmware allows it...
Yes, Cal scripts are now confidential.
As advised, return to Siglent in Germany for recalibration.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: KungFuJosh on October 16, 2023, 08:29:38 pm
Being on channel 1 (for example), press the "waveform" key then go to page 2, click on "DC", and with "DC Offset" you choose the voltage.
It leaves you from -10vdc to 10vdc

Try testing the ACVrms and see if that comes back accurately.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 16, 2023, 08:29:46 pm
ok thanks Tautech
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 16, 2023, 08:34:44 pm
SDG1032X - Sine wave 1KHz 1mVrms
SDM3055 = 0.991 mVrms

SDG1032X - Sine wave 1KHz 1Vrms
SDM3055 = 1.00850 Vrms
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: KungFuJosh on October 16, 2023, 08:39:40 pm
SDG1032X - Sine wave 1KHz 1mVrms
SDM3055 = 0.991 mVrms

SDG1032X - Sine wave 1KHz 1Vrms
SDM3055 = 1.00850 Vrms

That's pretty good. Now, try again, and view the DC voltage BEFORE applying any DC Offset.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 16, 2023, 08:46:33 pm
SDG1032X - Sine wave 1KHz 1mVrms
SDM3055 = 0.991 mVrms

SDG1032X - Sine wave 1KHz 1Vrms
SDM3055 = 1.00850 Vrms

That's pretty good. Now, try again, and view the DC voltage BEFORE applying any DC Offset.

When I talk about DC it is a pure DC, without AC of any kind.
and continues measuring the same as before
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: tautech on October 16, 2023, 08:48:22 pm
SDG1032X - Sine wave 1KHz 1mVrms
SDM3055 = 0.991 mVrms

SDG1032X - Sine wave 1KHz 1Vrms
SDM3055 = 1.00850 Vrms

That's pretty good. Now, try again, and view the DC voltage BEFORE applying any DC Offset.
Thinking same.
When compared to Output Characteristics from the datasheet (P8) it seems within spec.
https://int.siglent.com/u_file/download/22_06_23/SDG1000X_DataSheet_DS0201X_E01I.pdf
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: alm on October 16, 2023, 08:49:19 pm
Being on channel 1 (for example), press the "waveform" key then go to page 2, click on "DC", and with "DC Offset" you choose the voltage.
It leaves you from -10vdc to 10vdc
That's standard and documented in manuals for at least HP/Agilent/Keysight arbitrary waveform generators. Dave even made a video about it (https://youtu.be/XKI4DOduWs4).

Yes, Cal scripts are now confidential.
As advised, return to Siglent in Germany for recalibration.
I'm hoping Siglent know the difference between calibration and adjustment...

Why would they make adjustment scripts confidential? It's not like something like an AWG or power supply needs extremely expensive, rare equipment to adjust. For this adjustment a good DMM would probably all that's necessary.

When compared to Output Characteristics from the datasheet (P8) it seems within spec.
That's true, looking at the initial measurements they seem to be within the 1% + 3 mV uncertainty as specified in the manual. So I wouldn't count on sending it for adjustment changing anything.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: KungFuJosh on October 16, 2023, 08:49:48 pm
When I talk about DC it is a pure DC, without AC of any kind.
and continues measuring the same as before

An AWG is not a DC voltage reference.

For better reference, here's my SDM3055 and SDG2122X:

AWG: 1VACrms
DMM: 0.99997Vac
DMM: -2.748mVdc

AWG: 1Vdc Offset
DMM: 0.99922Vdc

AWG: 1mVdc Offset
DMM: -1.788mVdc

Here we can see that (ignoring likely minor voltage loss over the cables) that the DC offset is only an offset, and not an accurate voltage source without knowing what voltage you're actually offsetting.

Test DC before and after applying offset and you can see how accurate the offset is.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 16, 2023, 08:52:37 pm
Being on channel 1 (for example), press the "waveform" key then go to page 2, click on "DC", and with "DC Offset" you choose the voltage.
It leaves you from -10vdc to 10vdc

Try testing the ACVrms and see if that comes back accurately.

I already imagined it, but the calibration plus the round-trip shipping is respectable money.
What I don't understand is that Siglent itself has released a menu to calibrate them in the latest firmware of its multimeters, and in the SDG1032X it cannot be done.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: tautech on October 16, 2023, 08:53:20 pm
Yes, Cal scripts are now confidential.
As advised, return to Siglent in Germany for recalibration.
I'm hoping Siglent know the difference between calibration and adjustment...

Why would they make adjustment scripts confidential? It's not like something like an AWG or power supply needs extremely expensive, rare equipment to adjust. For this adjustment a good DMM would probably all that's necessary.
Checking if specs are met is outlined in the SM under Performance verification which anyone can do with the prescribed equipment.
Adjustment is a SW process and I know US branch has a Fluke 5200 for such needs. Not sure about EU.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 16, 2023, 08:54:48 pm
SDG1032X - Sine wave 1KHz 1mVrms
SDM3055 = 0.991 mVrms

SDG1032X - Sine wave 1KHz 1Vrms
SDM3055 = 1.00850 Vrms

That's pretty good. Now, try again, and view the DC voltage BEFORE applying any DC Offset.
Thinking same.
When compared to Output Characteristics from the datasheet (P8) it seems within spec.
https://int.siglent.com/u_file/download/22_06_23/SDG1000X_DataSheet_DS0201X_E01I.pdf

yes yes, but i need more precision :(
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: alm on October 16, 2023, 08:59:39 pm
Adjustment is a SW process and I know US branch has a Fluke 5200 for such needs. Not sure about EU.
Why would they need a seventies-era AC voltage source to adjust an arbitrary function generator? DMM, scope, maybe power meter and spectrum analyzer for flatness and harmonics, but why a voltage source?
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: KungFuJosh on October 16, 2023, 09:00:32 pm
yes yes, but i need more precion :(

You haven't demonstrated a lack of precision yet.

Measure your DC voltage WITHOUT offset applied.
Measure your DC voltage again WITH offset applied.
See if the difference in DC voltage is equal to what you set as the offset.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 16, 2023, 09:06:39 pm
yes yes, but i need more precion :(

You haven't demonstrated a lack of precision yet.

Measure your DC voltage WITHOUT offset applied.
Measure your DC voltage again WITH offset applied.
See if the difference in DC voltage is equal to what you set as the offset.

I'm using the SDG1032X as a precision  DC voltage source to test a 16-bit ADC to see how accurate and stable my design is, noise levels etc.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: KungFuJosh on October 16, 2023, 09:13:05 pm
I'm using the SDG1032X as a precision  DC voltage source to test a 16-bit ADC to see how accurate and stable my design is, noise levels etc.

As I said earlier, an AWG (in this case the SDG1032X) is NOT a precision DC voltage source. DC Offset is COMPENSATION only. The lower your voltage is, the less accurate it will be especially when you factor in voltage loss over cables. From what you posted, your AWG appears to be functioning as expected.

If you need a DC voltage source with that much accuracy, you need to buy or build an actual voltage source. This thread has a (not cheap) DIY DC voltage source: https://www.eevblog.com/forum/testgear/elektor-voltage-current-calibrator-project/new/#new (https://www.eevblog.com/forum/testgear/elektor-voltage-current-calibrator-project/new/#new)

There are some great inexpensive DC voltage sources out there, but not with super high accuracy or adjustable voltages. If you're on a budget, you'd probably be better off getting a calibrated DC Bench PSU with good adjustment ability.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: uargo on October 16, 2023, 09:20:54 pm
I'm using the SDG1032X as a precision  DC voltage source to test a 16-bit ADC to see how accurate and stable my design is, noise levels etc.

As I said earlier, an AWG (in this case the SDG1032X) is NOT a precision DC voltage source. DC Offset is COMPENSATION only. The lower your voltage is, the less accurate it will be especially when you factor in voltage loss over cables. From what you posted, your AWG appears to be functioning as expected.

If you need a DC voltage source with that much accuracy, you need to buy or build an actual voltage source. This thread has a (not cheap) DIY DC voltage source: https://www.eevblog.com/forum/testgear/elektor-voltage-current-calibrator-project/new/#new (https://www.eevblog.com/forum/testgear/elektor-voltage-current-calibrator-project/new/#new)

There are some great inexpensive DC voltage sources out there, but not with super high accuracy or adjustable voltages. If you're on a budget, you'd probably be better off getting a calibrated DC Bench PSU with good adjustment ability.

Yes I know, but it is what I have, I have the riden rd6006p that is quite accurate but it is switched and has some ripple.
I will have to buy a precise linear laboratory supply and use a potentiometer to reduce the voltage
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: KungFuJosh on October 16, 2023, 09:23:43 pm
Yes I know, but it is what I have, I have the riden rd6006p that is quite accurate but it is switched and has some ripple.
I will have to buy a precise linear laboratory supply and use a potentiometer to reduce the voltage

I get where you're coming from, but possession and budget don't redefine what a device is capable of. On the bright side, you don't need to waste any money on getting your AWG calibrated since it appears fine.

If I was doing something like what you're doing, I would look and see if the specs on something like the SPD3303X-E were sufficient for your needs. If not, then you need to look at voltage sources specifically, and that can get expensive for high accuracy.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: tautech on October 16, 2023, 09:37:13 pm
Adjustment is a SW process and I know US branch has a Fluke 5200 for such needs. Not sure about EU.
Why would they need a seventies-era AC voltage source to adjust an arbitrary function generator? DMM, scope, maybe power meter and spectrum analyzer for flatness and harmonics, but why a voltage source?
Sorry a $ $ Fluke universal calibrator, 5200 is wrong then.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: alm on October 16, 2023, 09:59:47 pm
I will have to buy a precise linear laboratory supply and use a potentiometer to reduce the voltage
I'd just build a fixed voltage divider, say 10:1 or 100:1, and adjust the voltage from the power supply/function generator. Be aware of the trade-off between output impedance / load regulation and dissipation in the resistors. If you use a pot, make sure you don't exceed their maximum dissipation.

If I was doing something like what you're doing, I would look and see if the specs on something like the SPD3303X-E were sufficient for your needs. If not, then you need to look at voltage sources specifically, and that can get expensive for high accuracy.
Is the SPD3303 really any better at the 1 mV level?

In addition to the Elektor project the PVDS2mini comes to my mind as reasonably affordable DC voltage source. Though not sure if Ian is currently able to produce them.

Sorry a $ $ Fluke universal calibrator, 5200 is wrong then.
Maybe I'm dense, but what good does a multifunction calibrator, which can generate a variety of DC and AC voltages (typically up to 1 MHz), currents and resistances do for adjusting a source like a function generator or power supply?
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: bdunham7 on October 16, 2023, 10:01:13 pm
As I said earlier, an AWG (in this case the SDG1032X) is NOT a precision DC voltage source. DC Offset is COMPENSATION only. The lower your voltage is, the less accurate it will be especially when you factor in voltage loss over cables. From what you posted, your AWG appears to be functioning as expected.

Not "precision" perhaps, at least not at the level the OP seems to need, but these AWGs are a perfectly fine source for DC voltages within reason and reading the specs, of course.  They're also not a terrible current source within their limitations.  They're stable enough to do calibration of lesser DMMs in tandem with a better one and supply enough current for many small DC breadboard projects.  You can even use them to test linear voltage regulator circuits for ripple rejection and transient response.  If you really want to be ridiculous, you can charge small batteries with them!
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: KungFuJosh on October 16, 2023, 10:34:54 pm
Not "precision" perhaps, at least not at the level the OP seems to need, but these AWGs are a perfectly fine source for DC voltages within reason and reading the specs, of course.  They're also not a terrible current source within their limitations.  They're stable enough to do calibration of lesser DMMs in tandem with a better one and supply enough current for many small DC breadboard projects.  You can even use them to test linear voltage regulator circuits for ripple rejection and transient response.  If you really want to be ridiculous, you can charge small batteries with them!

Precision was what he was asking for though. For basics it's fine. He wants under 1mV accuracy, but even the SDG2122X is spec'd as ±(1%+2mV) for DCV. TBH I've never used it for just DC, and it's pretty cool that the option is there.

Edit: I just checked and the SDG1032X is spec'd at ±(1%+3 mV) for DCV.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: bdunham7 on October 17, 2023, 12:21:30 am
Precision was what he was asking for though. For basics it's fine. He wants under 1mV accuracy, but even the SDG2122X is spec'd as ±(1%+2mV) for DCV.

Yes, he either needs an SMU or he needs to have a higher precision meter in parallel--and there still may be stability issues.

Interestingly, the SDG2042X appears to be the most accurate source of DC voltage I have on my bench, and that includes a PD 5020 "Precision DC Source" that sports an LM399 reference.
Title: Re: HOW TO CALIBRATE SIGLENT SDG1032X
Post by: KungFuJosh on October 17, 2023, 01:30:00 am
Interestingly, the SDG2042X appears to be the most accurate source of DC voltage I have on my bench, and that includes a PD 5020 "Precision DC Source" that sports an LM399 reference.

It's probably my most accurate (adjustable) voltage source on my bench too...at least until I finally bite the bullet and build that Elektor project. OTOH, I don't know if I should trust it, but I have a TP3005P bench PSU that my SDM3055s thinks is accurate. 2VDC shows 1.9999V, 20V shows 20.004V, 30V shows 30.005V, 10mV (lowest possible) shows 9.738mV. That's not too shabby.🤷