Products > Test Equipment
How to calibrate Siglent SPD1000X / SPD3303X / SPD3303X-E series power supplies?
<< < (22/27) > >>
Bad_Driver:
The SDM3055X behaves as expected! Very constant readings, in slow mode 5 Hz, in fast mode about 72 Hz!

So I'm pretty happy to spend 800 € for the SDM3065X with very weird and crappy acquisition behavior.  :palm:  |O

I try to compare your readings with the SDM3065 for the SDM threat, hope that Siglent takes care of this issue (Tautech where are you??).

I did a FFT with your slow readings, I assume this is not the SPD3303X??? (X = frequency/Hz)
oz2cpu:
>I did a FFT with your slow readings, I assume this is not the SPD3303X??? (X = frequency/Hz)

nope it is the oldie Oltronix_stabpac_mca_5-3 the most stable voltage i got :-)

I can make a slow-mode file SPD3303X tomorrow, just booted it up
I let the voltmeter stay on, all night too, this way they will be as stable as possible,
how long file do you think I need to record ?
only slow ? or also the other modes ?

Bad_Driver:
Plz try to make a file with 4096 readings. This is the limit for EXCEL FFT.
oz2cpu:
so this is the plan, agree ?

DC source : Siglent  SPD3303X set for 5.000 no load, been powered on for 24hrs at least.
slow 4096 points needed, 4096/5 = 819 sec, 13,6 mins
medium 4096 points needed, 4096/50 = 82 sec, 1,36 mins
fast 4096 points needed, 4096/92 = 44 sec, 0.74 mins

i add atleast a few secs, so we are sure you get over 4096, maybe it is a good idea to take double the length, and perform the test two times ?
see if results align ?

PS: how do i get the excel fft ? please email it to me, my email is my user name at gmail so that is easy to remember :-)
note : i have a hard time importing this CSV format, my excel dont accept the format.
Bad_Driver:
Follow this advice to get exact 4096 readings to usb. For best resolution 2 volts DC I think.
https://siglentna.com/operating-tip/sample-and-store-sdm3000/

For FFT I send you the information via mail tomorrow.

Edit: Please adopt the advice so, that datas are written direct to USB-disk, internal recordings will not have any time stamp.  :-//
Very old failure (I can‘t beliefe that this shall be a feature) in the firmware.
Navigation
Message Index
Next page
Previous page
There was an error while thanking