tv84:
--- Quote from: Gian2975 on May 19, 2019, 05:33:54 pm ---Could someone help me?
--- End quote ---
Send the SCPI command:
:SRLN <serial_number>
Tjuurko:
Сalibration method with SCPI command list:
http://old.siglentamerica.com/USA_website_2014/Documents/manual/SPD3000X_Service_Manual.pdf
Chapter 3 Calibration channel parameters
tautech:
Great, thank you.....I only checked the websites for this service manual a couple of days ago.
Edits need be made in earlier posts...........
ArthurDent:
I was looking at a photo of the back of the Siglent SPD3303X power supply and it looked vaguely familiar. Here is that photo next to a drawing from the manual of my slightly older GW Instek PSM-2010 power supply. I'd say that these were made by the same company. Not unusual to see a lot of rebranding on a lot of different types of test equipment.
tautech:
--- Quote from: ArthurDent on July 18, 2019, 05:34:26 pm ---I was looking at a photo of the back of the Siglent SPD3303X power supply and it looked vaguely familiar. Here is that photo next to a drawing from the manual of my slightly older GW Instek PSM-2010 power supply. I'd say that these were made by the same company. Not unusual to see a lot of rebranding on a lot of different types of test equipment.
--- End quote ---
For these 2 PSU's I doubt very much they're even closely similar except for a rear on view.
Quite a bit of Siglent's equipment is rebranded however off the top of my head I'm not aware of any of their PSU's.
Wanna do a 'I'll measure mine if you measure yours' ? :P
Actually there's no need to get that personal as here's the SPD3303X measurements: 225(W)×143(H)×278(D) mm