Author Topic: How to calibrate Siglent SPD1000X / SPD3303X / SPD3303X-E series power supplies?  (Read 4095 times)
Online ArthurDent
It probably isn't fair to compare the two power supplies because although I'm sure the Siglent is good, the PSM-2010 is, in my opinion, a far better supply which probably accounts for its price being about 3 times the Siglent.
Offline tautech
Sure, but this is not about mine is better than yours however checking the PSM-2010 specs I see its attractiveness for some. Can you series or parallel the 2 outputs ? Is each output isolated from the other ?
Avid Rabid Hobbyist
Offline chickenHeadKnob
  • doofus programus semi-retiredae
Arthur --
 They don't look the same at all. check positions of case screw holes (not including the fan mount as those are a standard)  and other details like the mains voltage switch which are design elements a rebrander would not change.
Offline kahuna0k
Ok, I've been able to calibrate voltage in both channels without problem, and now it is looking much better. But with current I have a problem. When I clear the calibration parameters for CH1, at 0.100mA I get no current at all through the electronic load, the multimeter is even measuring -0.0013mA. I'm able to adjust the 'b' parameter without problem and in fact at 2.5A everything matches to the mA, but when I start to go down in current, it starts to diverge, at 200mA I read 70mA and at 100mA I read 0. Any clue what should I do? I've tried several times, calibration only display current, only setting current and both values at the same time, with no change. I always start the calibration by clearing the values with
*CLCLS 2
*CLCLS 3
then set the display current values with the corresponding
CALibration:CURRent ch1,1,0.000   (I've also tried with -0.0013)
CALibration:CURRent ch1,2,2.3687 (or whatever I measure, usually after clearing the values I always get to 2.3687A)
and the same for setting current
CALibration:CURRent ch1,3,0.000
CALibration:CURRent ch1,4,2.3687
Any clue? Voltage calibration was so easy and effective, but I'm lost with this 0A at 0.100mA setting. I've also tried with a very small value, 0.0001 or something like that, just in case 0 is not a nice number for the firmware.
Thanks!
Offline kahuna0k
Answering to myself for the record. The points you choose for the calibration are arbitrary. The manual uses 0.1A and 2.5A for point 1 and 2 but you can use whatever, and the formula Y=a*X+b would be adjusted using the current value of X and the provided value of Y. I just used 0.2A and 2.5A and now the thing is precise to the mA. Great PSU! :D
The following users thanked this post: tautech, rmel, sjm
Offline sjm
From a later post the Service manual is now available with Cal procedures outlined:
http://old.siglentamerica.com/USA_website_2014/Documents/manual/SPD3000X_Service_Manual.pdf
Oh yeah, yesterday I was able to calibrate my (hacked) SPD3303X-E to have almost 1mA/1mV accuracy on all levels of output.
I used my bench DMM and now the requested, PSU self-measured and DMM measured voltages and currents are nicely lined up.
On some voltages, the PSU is almost comparable to some lower quality voltage references.   :P
The description of the calibration process in the manual is maybe a bit hard to read and not so logically written at times, but after some thinking and testing it out a few times, yeah it works.
It is important to keep the power output enabled while inputting the calibration commands!
Thanks a lot! Now I can sleep better!    ::)
BR, -sjm
« Last Edit: July 06, 2020, 07:29:29 am by sjm »
Offline tautech
A little something that is endorsed by the factory....SPD3303X/X-E units in series mode in series !
Avid Rabid Hobbyist
The following users thanked this post: tubularnut, BillB
Offline Elasia
Interesting, what do they say about two units in series but internally in parallel mode?