| how to connect sdg2000 to sds1000x hd for bode plot? |
| << < (3/3) |
| tautech:
--- Quote from: simmiv on August 23, 2024, 01:13:47 am ---Siglent has a document detailing the setup for Bode Plot using their SDG series Gens.. One issue I encountered is the fact that my Series 5000 Arb Gen (SDG 5082) doesn't have a "Channel Copy Coupling" ability whereby you can sync the two channels under the control of the Oscilloscope as detailed in the instructions.. --- End quote --- Does the channel coupling function not work when the scope takes control of the arb ? This was introduced in 5.01.01.15R1 FW in 2015. Have you the last FW version from 2018 installed ? https://int.siglent.com/upload_file/zip/firmware/Signal_generator/SDG5000_5.01.01.15R5_EN.zip |
| egonotto:
Hello, For USB connection, the USB Device from the generator must be connected to a USB Host from the scope. The supplied USB cable will therefore work. Best regards egonotto |