Author Topic: How to enable SIG ID (008) option for HP 856xE spectrum analyzer  (Read 1394 times)
Offline 1GhzTopic starter
  • Country: kr
WARNING: THIS WILL VOID YOUR WARRANTY ! ! ! (yes, really)
NOTE: Without the 298 Mhz SIG ID oscillator hardware, SIG ID AT MKR and SIG ID -> CF functions doesn't work. Only SIG ID function works.
Requirements
1. HP 856xE spectrum analyzer
2. Any GPIB adapter with command console software
3. 4mm HEX wrench
Procedure
1. Turn off the spectrum analyzer.
2. Remove the spectrum Analyzer cover.
3. Set the WR PROT/WR ENA jumper on the A2 controller assembly to the WR ENA position.
4. Turn on the spectrum analyzer.
5. Connect GPIB adapter to the spectrum analyzer.
6. Send the magic GPIB command to the spectrum analyzer.
Magic GPIB command: ZENTID 'HP<Model>,<Options delimited by comma>'
For example: ZENTID 'HP8560E,005,006,007,008'
7. Check the installed options from the spectrum analyzer. All is well, right? ;D
8. Turn off the spectrum analyzer.
9. Set the WR PROT/WR ENA jumper on the A2 controller assembly to the WR PROT position.
10. Secure the spectrum analyzer cover assembly.
11. ???
12. PROFIT!
Search keywords: 8560E 8561E 8562E 8563E 8564E 8565E SIG ID 008 Signal Ident Signal Identification
The following users thanked this post: dxl
Offline hafrse
Re: How to enable SIG ID (008) option for HP 856xE spectrum analyzer
« Reply #1 on: June 04, 2020, 10:06:28 am »
Hello,
Great information!
Regardng the 298 Mhz SIG ID Oscillator, is it a sperate module ? what is the part number of this part/module?
Thanks
  • Country: kr
« Reply #2 on: June 04, 2020, 10:19:50 am »
No, it's not a separate module. You need SIG ID enabled A15 RF board. Part number is 08563-60134.
Offline hafrse
« Reply #3 on: June 04, 2020, 04:14:32 pm »
Thanks again for the information, I assume there is no way to enable it on A15 by hardware modification except swaapping the board ?
Also, I need more information about the ZENTID command , suppose I have HP8562E with option 008 and I do a hardware uppgrade for some RF components (26.5Ghz RYTHM, attenuator,etc..) then executing ZENTID 'HP8563E' will change the model to HP8563E (-26.5Ghz)?  do I need to execute  ZENTID 'HP8563E'  or ZENTID 'HP8563E,008' ?
Thanks
  • Country: kr
« Reply #4 on: June 05, 2020, 03:37:45 am »
I didn't try it so I can't tell.
You need to execute ZENTID 'HP8563E,008'.
Offline hafrse
« Reply #5 on: June 14, 2020, 11:15:43 am »
Hello,
Will ZENTID 'HP8560E,005,006,007,008'  works also on 8560EC ? thanks
  • Country: kr
« Reply #6 on: June 14, 2020, 11:47:02 am »
I'm not sure.
But if the ID? GPIB command returns HP8560EC, I think you should execute ZENTID 'HP8560EC,005,006,007,008'.