I have a 2015 and a DMMcheck as well. Everything is within 0.02% other than AC current, which is 10% out.
This would seem to be out of spec:
ACCURACY1: ±(% of reading + % of range), 23°C ±5 °C
CURRENT CALIBRATION
3 Hz-10 Hz | 10 Hz–3 kHz | 3 kHz-5 kHz
-------------------------------+---------------+----------------
1.000000 A-1 ?A 0.30 + 0.04 | 0.10 + 0.04 | 0.14 + 0.04
3.00000 A-10 ?A 0.35 + 0.06 | 0.15 + 0.06 | 0.18 + 0.06
That said, it may be something to do with DMMcheck or it could also be a function of the probes/leads. I've had issues with AC current & DMMcheck with other meters.
Although it's unrelated and not in the same class, I noticed on my Uni-T U71D that in the specs, the mA range only goes to 400mA, then the 1A range takes over. Which means there is a hole between 400mA and 1A where nothing is measured. Perhaps the same thing is true with Keithley or there is a lack of accuracy in that range (e.g. it can't do sub .5A accurately), I haven't looked at the manual closely enough. What's clear is that if you want to measure things in the mA AC range, a 2015 is not the meter to get....
I'm not personally inclined to test any further since I have a Fluke 189 which is within 0.5% on AC current, but it is interesting that 3 different meters have the same issue. Amazingly, the Fluke is with 0.09% overall, which is really very, very good for a 10+ year old handheld.... No wonder they retain their value.
P.S. Attached is my test report. I do this for all my gear so I know what is most accurate....