This Hp 34401a is my resent eBay acquisition. When turned on all segments lit and the unit gives one beep followed by “ADDRS 23” error. After short delay it goes back to all segments lit and followed again by “ADDRS 23”. This ~1.5 seconds cycle goes forever.
When Self test is invoked; first starts with beep and "Testing.." then relays click, after short delay the unit displays “Test Fail”. The “Test Fail” message is followed by all segments lit and “ADDRS 23” in never ending cycle.
I checked voltages: both 5V and +/- 18V all within spec and clean. 15V is OK and -15v is a bit lower -14.37V. No overheating. All caps OK, except for one with slightly higher ESR. It has been replaced but this didn't make any difference. I red about possible display fault as not communicating with main PCB properly.
Serial number starts with 31 so I assume this unit was manufactured in 1991.
Anyone had this problem?? Any input greatly appreciated.