| Products > Test Equipment |
| HP 3455A strange problem around 14 - 80 V DC Input |
| (1/2) > >> |
| kblue:
Hello there! I rarely use Voltages above 10 Volts, so the discovery about this specific problem was kinda random (testing a DC power supply). With 50 VDC input, the meter jumps rather random between values like 22, 26, 34, 28 or sometimes nothing at all (blank). If i hit the 70+/- VDC mark, everything goes back to normal operation. Same with 14 Volts or below. So far nothing was changed within this specific meter (no caps, no nothing). Its rather untouched including the superb fan of unholy loudness :D Problem starts around 15 V input and stops at around 70 or 80 Volts DC ... between 15 and 80 it seems kinda "unstable". Sometimes 60 VDC input shows at 60 VDC, sometimes not. It seems to be within spec with 10 Volts input (compared to other meters 5-6 digit meters) or 100 VDC input. Mayhaps this small clip can help to "illustrate" the behaviour. I'm thinking about some dodgy relais or switches, dunno. Best wishes from germany! |
| Kleinstein:
A dodgy (e.g. oxidized contacts) relay is a possibilty. With 10 M input resistance the current is rather low and may not be enough to reach the wetting current of K5. Switching between the 10 V and 100 V range multiple times could be a 2st easy try to clean the contacts from usage. Is there a similar problem at around 200-600 V (1000 V range using a different divider setting) ? The test is running rather fast in the critical range. So it is hard to see if the test voltage is really stable, though the modern fluke meter looks a little more stable.It is still possible that the supply has a problem in that range and the 3455 is just faster to see the problem, maybe still assuming 60 Hz ?. |
| kblue:
Hello and thanks for the reply! I dont have a DC power supply that goes above 150 Volts. I tried another DC bench supply with the HP meter and the problem still occurs. [If i set the other (modern) DC power supply to 15.00 VDC the HP 3455A reads 14,9999 and is stable ... once i set the modern linear DC supply to 15.10 VDC the HP 3455A goes haywire. Going back to 15.00 VDC its normal again and reads accurate. So it seems to start with just above 15.00 volts DC. Everything here is running at 240 V 50 Hz. The "sample length switch" at the back of the 3455A is set to 50 Hz. Here is another video with slower increase in volts DC with some stops. Thanks in advance! |
| Kleinstein:
For a short time at the very top of the range the meter switched to the 1000 V range and was stable there. The time when the meter showed OL was a problem with autoranging. The wrong reading in the 100 V range made is switch between the 10 V and 100 V range rather often. This should already exercise the relay. So likely no luck there. One could try using manual 1000 V range and check the reading at some 20-50 V. If it is just the relay (K5) the reading should be unstable in the 1000 V range too, not just the 100 V range. To see if the problem is with the auto zero cycle one could try an AZ off mode (if available) and maybe a slower integration (should give less switching effect). Just for completeness a negative voltage could also give some insight. |
| kblue:
Hello again! Did some more tests ... Input 15.00 VDC from my modern DC power supply. Range Switch >10< shows 14.9999 VDC Range Switch >100< shows gibberish Range Switch >1000< shows gibberish 100 and 1000 are sometimes stable for a random amount of time. 10 and below is rock solid. Reversing the polarity is the same as above just ... well ... reversed. Decreasing the Sample Rate does not change the gibberish values with 100 or 1000 nor does AutoCal. So we could assume K5 is the culprit? As far as i know, these are no "ordinary relays". :/ Thanks for all your input so far! /edit: Did a bit more research and found this thread about "heating up" a reed-relay here on eevblog: https://www.eevblog.com/forum/repair/recovery-of-the-relay-k101(0490-1914)-in-hpagilent-34401a/ So i took my soldering iron and placed it at every wire/contact of K5 (3 on either side so 6 in total) and voila! Reading is fine but goes back to gibberish after some minutes of runtime. Now i need to source some of these relays i guess. They look kinda rare though :D |
| Navigation |
| Message Index |
| Next page |