Story behind the small wiggle meter on the hp 436A, 438A and others. When the hp 435 series of thermocouple based power meters were introduced the analog meter as typical and allowed observation in rates of change in power. Observing rates of change is poor to not possible with digital read out type displays. Here is a link to the hp journal introducing the 435 series of power meters. It is a LOT more complex and cleaver than it appears. An appreciation of how it worked needs to be studied to understand just how well the folks at hp addressed this design problem and solved it in a ver enduring way that lives on to this day.
http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1974-09.pdfRF power meters are much like a "Volt-Ohm-Amp" meter in the world of RF and microwave folks. Since most RF systems are 50 ohm based, this makes a plain and basic volt or current or current meter less useful as having a fixed impedance means power is more relevant than just volts, or current as either of those units are not relative to a impedance unless the impedance is known. This is why power based on a 50 ohm system is used as a normalization for measuring power in RF and power systems.
With the introduction of the hp 435 & hp 8481, power measurements were made easier. Before this system, thermistors/thermistor mounts and all the problems associated with that system of measurement was a source of never ending niggles. The analog meter work well, it was accurate, allowed rate of change observation and was reliable, durable and accurate.
In time the folks at hp decided to update aka "improve" the 435 series of meters by offering a digital readout version known as the 436A. During instrument trials, users trying to observe rate of change (in power) discovered it was near impossible with the digital display. To address this real world instrument requirement, the folks at hp added a small wiggle meter specifically designed and intended to display rate of change.
In the home lab, the 435A/B. 436A, 438A (fan sounds like a jet taking off) all live together as the complements to each other and allow great power measurement flexibility. The power meters are common, the power sensors and cables are precious as they are compatible over a broad range of hp power meters. These can be amazingly accurate for years and years. The power sensor heads can be fragile and easily damaged if excessive power is applied. They are tolerant to this kind of abuse to a limit. Go past that limit and the power sensor is dead.
Wiggle meters have their place and offer advantages digital displays cannot. Another case of knowing how to apply what work best for a given measurement requirement.
In the passing decades since the introduction of hp's thermocouple based power sensor, came the diode based power sensor and eventually other manufactures got into this game including Gigatronis (previously Systron Donner), Wiltron / Anritsu and others including made in China knock-offs.
Bernice