Author Topic: HP 54503A scope, address error  (Read 1285 times)
Offline iblackfordTopic starter
« on: November 10, 2022, 02:12:16 am »
Hi All,
I have recently acquired an HP 54503A 500Mhz scope. I thought this scope had the standard NVRAM issue, so I replaced it, however the problem still exists.
I've attached the picture of this error...I can't find anything in the service manual to isolate this error. Any ideas?
Thanks, Ivan
Re: HP 54503A scope, address error
« Reply #1 on: November 10, 2022, 01:44:19 pm »
I should add that I _did_ try numerous attempts at holding down a button on startup to attempt a calibration, and nothing was working. I also made sure that the calibration switch was set to unprotected while attempting a calibration.
Could this be a ROM/RAM issue? does anyone know where I can find firmware dumps for this or similar models? Schematics?
Online nctnico
« Reply #2 on: November 10, 2022, 01:51:03 pm »
I'd start here. Looks like the 54501A model is related so likely many, if not all information applies to your oscilloscope as well. HP / Agilent has a habit of basing various oscilloscope models on the same hardware platform with only the last digits of the model number being different.
http://www.ko4bb.com/getsimple/index.php?id=manuals&dir=HP_Agilent
Online Miek
« Reply #3 on: November 10, 2022, 04:15:34 pm »
The "Address Error" a low-level processor exception meaning that the firmware tried to access memory at an odd-numbered address. I would expect the only cause of that to be some corruption happening somewhere, so yeah, I'd suspect bad ROM or RAM.
« Reply #4 on: November 10, 2022, 04:39:57 pm »
Cool...thanks, that's what I thought it might be.
I'm curious, how did you know that it was an odd address? Is there any further debug information that I can glean from the various numbers on the screen?
Thanks, Ivan
Online Miek
« Reply #5 on: November 10, 2022, 05:06:58 pm »
Just from bits of documentation on 68k series processors:
An address error exception occurs when the processor
attempts to read a 16-bit word or a 32-bit Iongword at an
odd address. Attempting to read a word at an odd address
would require two accesses to memory - - one to access
the odd byte of an operand and the other to access the
even byte at the next address.
from
http://marc.retronik.fr/motorola/68K/68000/Exception_handling_in_the_68000_[A.CLEMENT_1986_19p].pdf
I was pretty sure it's a 68k since most of HP's instruments from that era had one, and looking at the 54501a firmware it has strings for the other standard 68k exceptions (see attachment).
It should be possible to glean more information from the numbers on screen (they show the state of various CPU registers at the time of the exception), but it'd require disassembling & reverse-engineering the firmware image.
« Last Edit: November 10, 2022, 05:09:58 pm by Miek »
« Reply #6 on: November 10, 2022, 05:29:13 pm »
Thanks for all the help, this is making a lot more sense now. Reading the service manual, I can confirm that it is indeed a 68000. I have a coworker that has done a bunch of bare metal 68k coding in the past, so he may be able to disassemble this and help me out a bit.
However, I'm going to start reseating all of the ROM chips first, as they are socketed, and this is something free to try. :)
This scope has sat on a shelf for a couple of years neglected, but I don't remember it having the address error when I diagnosed it back then. I'm hoping that it's not any of the socketed ram chips.
Ivan
Online Miek
« Reply #7 on: November 10, 2022, 05:38:06 pm »
No problem :-+
If you're able to dump the data from the ROMs that would help - they should have a checksum or CRC to test for any corruption in the firmware.
« Reply #8 on: November 10, 2022, 10:41:42 pm »
I removed and reseated all the ROMs...and..... :)
Thanks for all the help!
Online nctnico
« Reply #9 on: November 10, 2022, 11:40:07 pm »
Cool! Contact problems are the first to look at and typically I start by reseating all chips & connectors. Sorry for not mentioning it.  ;D
« Reply #10 on: November 11, 2022, 01:39:48 am »
It's on my list of things to check as well, I just figured given that it's such an old and popular unit, maybe there are some folks out there who have experienced corrupt roms or some other age related failure.
FYI...if anyone needs a replacement for a DS1230/DS1235 let me know, I rolled my own: