Looks about as expected overall, though I was surprised to see the simultaneous display of both DUTs, that's pretty cool. I couldn't really see, is there much control over that feature as far as positioning/switching rate? Any noticeable effect on measurements in that mode?
As far as the digitising thing, the easiest way I found for the 576 was tapping off the deflection plates via a little adaptor circuit I built up to attenuate and offset. I tried a couple other things but doing it this way meant that I can move stuff around with the curve tracer display controls and the digitised version will always match what I see on the CRT. The only catch is that you need to tell the software what the range is any time you change vertical/horizontal settings as the digitised signal is always just a scaled display (in other words, the digitiser never sees a 0-1V or 0-1000V or 0-500mA signal, it sees whatever voltage it takes to drive the CRT from 0-10 divisions regardless of your test voltages/currents). Obviously if you could be assed you could set up auto range switching with a bunch more wiring and code, but so far I just save templates with the correct ranges for parts I test the most.