Author Topic: HP 8153A Lightwave multimeter - does not power up  (Read 2493 times)
Offline asilva10Topic starter
« on: October 26, 2018, 03:46:19 pm »
I have a HP8153A that does not power up. I was not able to find any schematics or service manual for this instrument.
I made quick measurements at the power-supply board (HP P/N 08153-66503) and it seems that the input stage is okay and after the rectification I have +165V (input 110Vac). But after the switching stage, no voltage level. I cannot detect any sign of "oscillator on". I believe that the driver is a component Q101, but I am not sure.
I tried to look at eBay and found a lot of these instruments with the same problem (no power). Any hint?
Offline Jwalling
  • This is work?
Re: HP 8153A Lightwave multimeter - does not power up
« Reply #1 on: October 26, 2018, 03:55:37 pm »
There's a 8 pin DIP that controls the switcher, I forget the part #. Replace it and you'll be good to go. I've fixed at least 10 of these in the past where that chip was bad.
Jay
System error. Strike any user to continue.
« Reply #2 on: October 26, 2018, 04:09:12 pm »
@Jwalling - thank you!! Yes, I found: TDA4605 - a switching controller - it must be this guy!!
I will let you know. Thanks!
« Reply #3 on: November 02, 2018, 05:55:27 pm »
Problem solved: put a 8-pin DIP socket and installed a new TDA4605. Everything working now !!
Offline Jwalling
  • This is work?
« Reply #4 on: November 03, 2018, 09:48:02 am »
For some reason, the chips made in Austria have a high failure rate.
Jay
The following users thanked this post: K9DTV
Offline K9DTV
« Reply #5 on: May 14, 2021, 11:10:13 pm »
I picked up a non-working HP 8153A with plugins and went looking for schematics, found nothing!
Ended up reading this post and ordered the TDA 4605 after reading this threat.
Got the chips in today and unit is up and running, Thanks for your post on this, and I made sure the parts where not made in Austria.
Offline Jwalling
  • This is work?
« Reply #6 on: May 15, 2021, 08:15:30 am »
You're welcome. There's also an NTE equivalent part that works. Part is NTE7154
Jay
Offline picburner
« Reply #7 on: June 10, 2021, 06:33:45 am »
If it can be useful in the attachment there are the power supply's schematics.
Offline rammer
« Reply #8 on: October 05, 2023, 07:45:00 pm »
Hello,
I had the same problem, the mainframe was dead. I have replaced the switching IC and after that the power supply looks OK (correct voltages). The 8153A starts but the keys do not work and an inserted source can not be enabled.
The mainframe did work prior to the issue flawlessly.
Any ideas?