I have a HP8153A that does not power up. I was not able to find any schematics or service manual for this instrument.
I made quick measurements at the power-supply board (HP P/N 08153-66503) and it seems that the input stage is okay and after the rectification I have +165V (input 110Vac). But after the switching stage, no voltage level. I cannot detect any sign of "oscillator on". I believe that the driver is a component Q101, but I am not sure.
I tried to look at eBay and found a lot of these instruments with the same problem (no power). Any hint?
Jwalling:
There's a 8 pin DIP that controls the switcher, I forget the part #. Replace it and you'll be good to go. I've fixed at least 10 of these in the past where that chip was bad.
asilva10:
@Jwalling - thank you!! Yes, I found: TDA4605 - a switching controller - it must be this guy!!
I will let you know. Thanks!
asilva10:
Problem solved: put a 8-pin DIP socket and installed a new TDA4605. Everything working now !!
Jwalling:
--- Quote from: asilva10 on November 02, 2018, 05:55:27 pm ---Problem solved: put a 8-pin DIP socket and installed a new TDA4605. Everything working now !!
--- End quote ---
For some reason, the chips made in Austria have a high failure rate.