Gandalf_Sr:
I was looking at one of these on eBay the other day - am I right in saying that the tracking generator option only goes up to 2.9 GHz?  Any other models I should consider?
MarkL:

--- Quote from: Gandalf_Sr on September 30, 2015, 04:23:04 pm ---I was looking at one of these on eBay the other day - am I right in saying that the tracking generator option only goes up to 2.9 GHz?  Any other models I should consider?

--- End quote ---

You are correct.

I see you've already started a separate thread for other model suggestions.
kada:
Thank you all for helping. After the YTO replacing the analyzer works fine. FREQ/AMPLITUDE calibration passes.
I also have another 8595E with defective YTF  :palm: But it works fine at the band B0 only (9 kHz - 2.9 GHz).
MarkL:
Glad you got it working!

I also have two 8595E and one has the exact same problem that the YTF is dead, leaving band 0 still working.

Thinking I couldn't make matters worse, I opened the YTF and one of the YIG sphere's rolled away.  Doh!  It had come off its mount.  That's the end of that.

"Fixed" it with a sticker on the front panel reminding me band 1 is dead.
AF6LJ:

--- Quote from: kada on October 09, 2015, 02:35:03 pm ---Thank you all for helping. After the YTO replacing the analyzer works fine. FREQ/AMPLITUDE calibration passes.
I also have another 8595E with defective YTF  :palm: But it works fine at the band B0 only (9 kHz - 2.9 GHz).

--- End quote ---
Glad to hear you got it working Great. :)  :-+