Hydron:

--- Quote from: mawyatt on January 18, 2023, 05:38:57 pm ---We have an old HP Rev 05-01-01 and a AG Rev 11-05-02 (our first 6-1/2 DMMs), but neither has anything connected to the RS232 nor HP-IP ports. What would folks recommend as gaining remote access with an adapter, from Ethernet (preferred), with an intent of looking into these "enabling" features?

--- End quote ---
The 05-01-01 unit will not be able to use the new features (if you look in the FW the menu strings are absent), the 11-05-02 will work fine though (I have the same FW).
I'm not sure what the best ethernet options are (GPIB-Ethernet adapters are rather pricey), but a simple USB-RS232 cable plus a null modem adapter (must have TX/RX and DTR/DSR crossovers) will work fine if you go and turn on the appropriate options in the IO menu. You won't get maximum reading rate via RS232, but that's not an issue in the high-res modes.
iMo:

--- Quote from: Hydron on January 18, 2023, 07:15:41 pm ---
--- Quote from: mawyatt on January 18, 2023, 05:38:57 pm ---We have an old HP Rev 05-01-01 and a AG Rev 11-05-02 (our first 6-1/2 DMMs), but neither has anything connected to the RS232 nor HP-IP ports. What would folks recommend as gaining remote access with an adapter, from Ethernet (preferred), with an intent of looking into these "enabling" features?

--- End quote ---
The 05-01-01 unit will not be able to use the new features (if you look in the FW the menu strings are absent), the 11-05-02 will work fine though (I have the same FW).
I'm not sure what the best ethernet options are (GPIB-Ethernet adapters are rather pricey), but a simple USB-RS232 cable plus a null modem adapter (must have TX/RX and DTR/DSR crossovers) will work fine if you go and turn on the appropriate options in the IO menu. You won't get maximum reading rate via RS232, but that's not an issue in the high-res modes.

--- End quote ---
In order to get a continuous stream of results via rs232 you have to enable the "talk only" mode in the HP-IB settings (do set the address 31 via front buttons). With the 9600 8N2 (you have to set in the meter too via front buttons) it works with some lower NPLCs as well. Simply open TeraTerm or other terminal app and you will see the values coming in..
Hydron:
If you're okay with polling using READ? (or setting up triggers and using INIT/FETCH) then you don't need to use the "talk only" mode and can adjust settings via SCPI as well. I suspect for most remote controlled operations using SCPI will be desirable. There is also a driver for HKJ's "TestController" software that can be used for graphing, scripting etc (though it needs updates to take advantage of the new TEMPerature measurement functions etc).
Hydron:
I've done a little testing of what new SCPI commands are available with the "upgraded" meters, here is what I came up with:

--- Code: ---Commands with ? as a suffix are read only, all others are read/write unless noted (using ? and no arguments for read). As usual only capitalised letters are required, syntax is the same as used by HPAK in their manuals.

Math operations:
Scale math operations (only one math operation available at a time, e.g. no min/max simultaneous with SCALE):
CALCulate:FUNCtion SCALe
CALCulate:SCALe:GAIN
CALCulate:SCALe:OFFSet

Average math operation (in addition to existing):
CALCulate:AVERage:SDEViation?
CALCulate:AVERage:PTPeak?


Temperature measurment:
DIAGnostic:TEMPerature? <- read internal on-chip temperature sensor (I assume in U101-B), note that it operates relay(s) when reading it for some reason regardless of selected function

CONF? <- when in TEMP mode also returns transducer type and subtype in addition to range/resolution (range fixed as "1")

CONF:TEMPerature {TCouple|RTD|FRTD|THERmistor|DEF},{TC type, or (F)RTD alpha, or Thermistor type},{<range>|MIN|MAX|DEF},{<resolution>|MIN|MAX|DEF} <- Temperature measurement mode/transducer/subtype; no query possible, transducer DEF(ault) = last used, range is ignored

UNIT:TEMPerature {Cel|Far|K}

[SENSe]:TEMPerature:TRANsducer:TYPE {TCouple|RTD|FRTD|THERmistor|DEF} <- FRTD is 4W RTD, RTD is 2W, DEF(ault) = TC (unlike for CONF:TEMP)

[SENSe]:TEMPerature:NPLCycles {0.02|0.2|1|10|100|MINimum|MAXimum}

Thermocouple options:
[SENSe]:TEMPerature:TRANsducer:TCouple:TYPE {B|E|J|K|N|R|S|T} <- TC type
[SENSe]:TEMPerature:TRANsducer:TCouple:RJUNction {<temperature>|MIN|MAX} <- Reference junction temperature for TC, -1 to 55C range

RTD options (shared between 2W RTD and 4W RTD, either RTD or FRTD will work in SCPI command):
[SENSe]:TEMPerature:TRANsducer:RTD:TYPE {85|91} <- RTD alpha value, 0.00385 or 0.00391
[SENSe]:TEMPerature:TRANsducer:RTD:RESistance:REFerence {<R0 resistance>|MIN|MAX} <- RTD reference (25C) resistance, 4.9 - 2100 ohm range

Thermistor subtype option:
[SENSe]:TEMPerature:TRANsducer:THERMistor:TYPE {2200,5000,10000} <- Thermistor type (in ohms)


Custom aperture:
<measurement function>:APER {<aperture value>|MIN|MAX} <- On a normal 34401A this will only accept values which correspont to the standard NPLC settings. When custom aperture option is enabled, it will instead accept values anywhere in the range of 0.4ms to 2s. The NPLC option will not accept "non-standard" values - when set to a non-standard value the next highest standard value will be used instead.
--- End code ---

I haven't managed to work out what (if any) SCPI commands control settings save/recall functionality (not critical for remote use tbh), and the custom aperture settings work a little differently compared to when set from the front-panel:
- custom apertures set manually via the front panel menu system (by selecting "custom" in the resolution menu, enabling the extra aperture time menu) will always show full 6.5digit resolution regardless of aperture time setting (which has a range of 0.4ms to 2s)
- There is no explicit "custom aperture" setting via SCPI that I found, only that you can use the <function>:APER SCPI command set it to arbitrary values (in the same 0.4ms to 2s range) after the POKE 32 upgrade
- any SCPI-set aperture that corresponds to a "normal" NPLC setting will result in the meter going back to showing the "normal" resolution, e.g. 5.5 digits for a 50Hz aperture of 20ms (1 NPLC)
- again this is probably not a big deal for remote control, as the resolution returned by SCPI behaves as it does normally (i.e. is independent of aperture time/NPLC and greater than display resolution)

I'll also be uploading a modified instrument definition for HKJ's SW to enable the Temperature functions here shortly: https://www.eevblog.com/forum/testgear/program-that-can-log-from-many-multimeters/

Edit: Seems I spoke too soon about save/recall SCPI commands. I took a look at the 34410A manual and the following seem to work:

--- Code: ---*RCL 0 (or another number, only 1 slot it seems) <- recall saved state
*SAV 0 (or another number as above) <- store current state
MEMory:STATe:RECall:AUTO {OFF|0|ON|1} <- turn power-up recall on/off
--- End code ---
The other 34410A memory commands I tried didn't seem to work.
Kosmic:
Most of the hidden SCPI functions are documented in the manual of the HP 34970A (see attachment).

I guess the custom aperture can be set with this:

--- Quote ---[SENSe:]
 VOLTage:DC:APERture {<time>|MIN|MAX}[,(@<ch_list>)]
 VOLTage:DC:APERture? [{(@<ch_list>)|MIN|MAX}]
--- End quote ---