Recently acquired an HP E2050A off eBay, but can't seem to have the device reset to default settings. 

Have tried combinations / lengths of holding the 'Config Preset' button while powering on but doesn't seem to have any effect.  Based on the manual, expect to see the 'Conn' LED flash after this but no such luck. 

GPIB | SRQ LED remains on all the time which doesn't seem to be expected.  Fault LED doesn't come up either.  Anyone run into this?
 

Offline Algoma

  • Frequent Contributor
  • **
  • Posts: 291
  • Country: ca
Re: HP E2050A Gateway | Can't Reset to Default
« Reply #1 on: April 22, 2021, 01:08:33 am »
https://www.eevblog.com/forum/testgear/hp-e2050a-mini-teardown/

gslick responds with some details on how to reset the unit from the console:
Quote
You can connect a terminal or PC running a terminal emulator to the E2050A RS-232 port using a null modem cable at 19200,N,8,1 and after hitting <CR> it should bring up the configuration menu.
« Last Edit: April 22, 2021, 07:58:43 pm by Algoma »
 

Offline Venturi962Topic starter

  • Regular Contributor
  • *
  • Posts: 123
  • Country: us
Re: HP E2050A Gateway | Can't Reset to Default
« Reply #2 on: April 22, 2021, 02:30:38 am »
Thanks - did see that and no result either.  Don't quite understand why the SRQ LED is on without any device plugged in as I believe that is reserved for devices.  Also seems that SRQ is actually idle, measures 5V on the GPIB Controller (SMJ9914A).
 

Offline gslick

  • Frequent Contributor
  • **
  • Posts: 580
  • Country: us
Re: HP E2050A Gateway | Can't Reset to Default
« Reply #3 on: April 22, 2021, 06:02:26 pm »
During power up of my E2050A the front panel LEDs briefly turn on in a few different patterns. After less than 10 seconds with nothing connected to the HP-IB / GPIB and LAN connectors the only LED that remains on is the Power LED.

If the SRQ LED remains on with your E2050A with nothing connected to the HP-IB / GPIB connector and you cannot get any response on the serial port with a null modem cable connection at 19200 baud, 8 bit, no parity, then maybe your E2050A has a hardware fault. If you just acquired it off of eBay, it might not be too surprising if it turns out to be dead.

Serial port output at 19200,N,8,1 after sending a CR to the E2050A after power up:

Code: [Select]
Welcome to the E2050A LAN/HP-IB Gateway configuration utility.

Commands
  ?                Show additional commands
  exit, quit       Exit WITHOUT saving configuration changes
  reboot           Save configuration changes and restart
  status           Show the LAN/HP-IB Gateway connection status

Configuration Parameters
  hostname:        E2050           # Internet domain name
  hardware-addr:   0800091AEE6E    # Ethernet station address
  ip:              192.168.1.20    # Internet Protocol address
  default-gw:      192.168.1.1     # Default subnet gateway IP address
  subnet-mask:     255.255.255.0   # Network subnet mask
  syslog-svr:      0.0.0.0         # Syslog server IP address
  bootp:           OFF             # Obtain config via BOOTP/TFTP
  lan-timeout:     0               # LAN connect timeout in seconds
  io-timeout:      120             # Server I/O timeout in seconds
  allow:           *               # IP allow list

  hpib-address:    21              # HP-IB System Controller Address
  hpib-name:       gpib0           # HP-IB interface symbolic name
  hpib-unit:       7               # HP-IB logical unit number
> ?
Available commands are:
  allow: <ip> [<ip>]       Set the allow list for IP security
  bootp: <ON|OFF>          Turn ON or OFF use of BOOTP/TFTP
  bye                      Quit without saving configuration - see reboot
  config                   Show current settings
  default-gw: <ip address> Set the default subnet gateway address
  exit                     Exit without saving configuration - see reboot
  factory-reset            Reset configuration to factory defaults and reboot
  hardware-addr            Display the ethernet station address
  help                     Print help information
  hostname: <string*35>    Set the internet domain name
  hpib-address: <0-30>     Set the HP-IB system controller address
  hpib-name: <string*15>   Set the HP-IB interface symbolic name
  hpib-unit: <1-255>       Set the HP-IB logical unit number
  io-timeout: <seconds>    Set the server I/O timeout in seconds.  0 is Off.
  ip: <ip address>         Set the IP address
  lan-timeout: <seconds>   Set the LAN connect timeout in seconds.  0 is Off.
  quit                     Quit without saving configuration - see reboot
  reboot                   Save configuration and restart
  status                   Show the LAN/HP-IB Gateway connection status
  subnet-mask: <mask>      Set the network subnet mask
  syslog-clear             Clear the syslog
  syslog-display           Display the syslog contents
  syslog-svr: <ip address> Set the syslog server address
  version                  Show the firmware revision
> version
Software revision A.01.03
Bootrom revision A.01.00
« Last Edit: April 22, 2021, 06:04:20 pm by gslick »
 

Offline gslick

  • Frequent Contributor
  • **
  • Posts: 580
  • Country: us
Re: HP E2050A Gateway | Can't Reset to Default
« Reply #4 on: April 22, 2021, 06:16:55 pm »
If you bought this one, unfortunately you might be out of luck:

https://www.ebay.com/itm/164773841521

Quote
This lot is for one HP E2050A GPIB HPIB Over IP Ethernet LAN Internet Converter Gateway.

On power up, LAN Tx and Rx lights don't activate and HP-IB SRQ light stays light, yellow. System function is unknown, offered here for parts.

This unit will arrive in the condition shown. It is being sold AS-IS and WITHOUT WARRANTY.
 

Offline Venturi962Topic starter

  • Regular Contributor
  • *
  • Posts: 123
  • Country: us
Re: HP E2050A Gateway | Can't Reset to Default
« Reply #5 on: April 22, 2021, 07:21:05 pm »
Thanks gslick - good thing I didn't pay seller's full price.  Looks like I got a paperweight.
 

Offline Tomorokoshi

  • Super Contributor
  • ***
  • Posts: 1212
  • Country: us
Re: HP E2050A Gateway | Can't Reset to Default
« Reply #6 on: March 23, 2022, 01:18:40 am »
I have the same problem with one E2050A. I have another that works. Here is the timing of the indicator light sequence:

TimeWorkingNot Working
0.0Power OnPower On
1.0Power, Fault, Conn, Talk, Listen, SRQPower, Fault, Conn, Talk, Listen, SRQ
2.0Power, SRQPower, SRQ
2.1Power, Listen, SRQPower, SRQ
2.2Power, ListenPower, SRQ
3.0Power, Talk, Listen, SRQPower, SRQ
3.1Power, Talk, SRQPower, SRQ
3.2PowerPower, SRQ
4.5Power, Tx, RxPower, SRQ
4.6PowerPower, SRQ
4.7Power, Tx, RxPower, SRQ
4.8PowerPower, SRQ

It seems that from around 1 second after power on to around 3 seconds it is testing the GPIB controller. The failed unit stops somewhere in the SRQ part of the test. Probing on the 68360 processor, it seems to get a /BERR (Bus Error). Other lines have activity consistent with /BERR being asserted.
 