I was given a Fluke 87 (original) as it was not powering up at all. At first it looked easy. Q8 was faulty, once i replaced that from a donor unit it did power up. However, it did only get to the point when all segments of the display is lit, and stayed that way.
I looked at the usual suspect, such as the mode selector thick film contacts, cleaned everything again, etc. but to no avail.
I noticed that if I wiggle the mode selector, in some in-between positions, the unit powers up in AC V mode, which is the first mode next to off.
I could not get any further.
I believe, although I did not yet check it with a scope that it figures out the mode selection by varying the output on AP6 (as it seems to be an analog out port) and getting the input on P03 (digital input), which has a fixed threshold. Depending on where the mode selector is, P03 will flip at different values of AP06. Based on this, I suspect that U4 is faulty, as there doesn not seem to be any other components that can fail to produce this result. This is why the only mode is AC V as p03 will only turn on if the mode selector is in a between setting.
I will need to check if my theory holds and scope on AP06 and P03. The contacts and the divider chain of the 10k resistors all look ok.
Or do you have any idea on what else can be wrong?