Products > Test Equipment

HP3455A DVM repair

(1/2) > >>

intabits:
I finally got around to working on this HP3455A DVM, which I got for cheap at a hamfest because the seller said it emitted smoke when he turned it on that same morning. I did a teardown video of it back in 2018, and at that time could not find any obvious source of smoke.

The fault was showing "0L" on startup, and the self-test produced errors on cal constant tests 11 to 0, except 7 and 4. The power supplies were all good, and no ripple was evident using a scope.

I made a "service cable" and used my good 3455 to do the "half-split" procedure, which connects the inguard of one unit to the outguard of the other. This identified the inguard section as being at fault.

Also, the 10V reference voltage was about 7V and not steady (jumping about I assume due to auto-cal switching). I removed the reference module and powered it up stand-alone, and it produced 10V. So something was loading it down.

The service manual has troubleshooting flowcharts specifically for each of the cal constant test error codes, but explicitly says not to use self-test mode to get into a particular cal constant test state for troubleshooting. It gives a procedure to select one of these states without using test mode, but I couldn't get that to work, on either of my 3455s.

So I posted a query in the Repair section of the forum asking how to do this cal state selection. I didn't get an answer on that, but Kleinstein posted a response that made me realize that my understanding of the JFET operation in the input switching section was all screwed up. After some study, I had a much better grasp on how the switching operated.

I tried using the flowchart for error state 11, but abandoned it because the voltages I measured were a bit off from what the flowcharts specified, which I blamed on the bad 10V reference voltage.

Trying another tack, and dreading a bad JFET, which my research showed would be hard and/or expensive to replace, I pinned my hopes on a bad comparator, which apparently is a common fault.

So I set about measuring voltages on the inputs and outputs of all the comparators (LM339) that control the JFETs. I found three were not pulling the JFET gates to -24V when their +input was 0V (logic 0 from the inguard controller) while their -input was at the 1.2V of the logic reference rail. All three were in A10U9. (The 4th element of U9 was used elsewhere and I didn't test it)     

After replacing U9 (socketed), the unit came good!
A few crude tests showed it agreed closely with the good 3455 on DCV, ACV and Ohms. I was a happy chappy!


factory:
The smoke could have been from a failed RIFA metallized paper capacitor (or a Schaffner line filter), these are often found in HP stuff of this age, they are usually fitted to the mains inlet (but not always), sometimes quite well hidden too.

David

Kleinstein:
The 2.9 V level for one of the U4 comparators looks OK: 3.5 V is about right for a loaded TTL chip and the diode looses another 0.6 V or so.  The comparator gets a reference level of 1.2 V, so that 2.9 V are fine there.

A bad LM339 is for some reason relatively common, not just with this meter. Not really sure why - could be just aging (e.g. purple plague ?) and some bad series.  The good thing is that they are readily available.

intabits:
That's a good point that I hadn't thought of. (maybe because the seller had indicated smoke coming from the opposite side of the unit).
I've just removed the mains safety cover for the first time, but there were no RIFA's or any visible damage in there.

Kleinstein:
The mains filter cap is indeed the main suspect for smoke. It does not have to be a Rifa, other brand capacitors also fail by releasing smoke.  If inside a line filter the damage may not be that visible. Another part is quitestion would be a MOV to protect from overvoltage at the mains input, so usually close to the filter capacitor.
The 3rd suspect for smoke would be a tantalum capacitor in drop shape at a supply with enough power for ignition.

Navigation

[0] Message Index

[#] Next page

There was an error while thanking
Thanking...
Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod