I have four HP5335a and my concern centers around triggering.
case 1
Say I have 10MHz on input A. Trigger in AUTO mode, LED flashes on both A and B
Is this the way supposed to be? There is nothing connected on B port.
case 2
Say I have 10MHz on input A. Trigger in AUTO mode LED flashes ONLY on port A. B is either not lit or steadily lit. It never flashes.
Oddly enough I have case 1 and case 2, two each.
By the way, if I connect signal to port B with nothing on A, 3 out of 4 has A and B flashing. Last one is either blank or steady depending on setting of trigger.
It sort of looks like cross talk. If I enable 50 ohm input, only trigger LED that flickers is the one input is connected.
It WAS a cross talk. When I reduced the input signal from 1Vp-p to 100mVp-p, the cross talk went away.