Products > Test Equipment
HP8175A Digital Signal Generator v2.2 FW & Manuals / Datasheets
(1/1)
garrettm:
A forum member asked me to upload a dump of my HP 8175A FW and I figured I would upload that along with .pdf copies of the manuals and other product literature I've collected for anyone that is interested.

-- Attached Files --

Firmware:

HP 8175A Digital Signal Generator v2.2 FW
UV EPROM: AM27C256 x6

PDFs:

1989 Datasheet from HP Catalog
1992 Datasheet from HP Catalog
Page 3 of Sep/Oct 1985 issue of Measurement Computation News

-- Files Too Large to Attach --

HPJ Apr 1987 Article (pgs 4-20)
HP 8175A (Opt 001) Operating and Programming Manual
HP 8175A (Opt 001) Service Manual
HP 8175A (Opt 002) Operating & Programming Manual
HP AN341-1 - HP 8175A / HP 1630G Stimulus-Response System
HP AN341-2 - Simulating Sensor Signals

https://www.mediafire.com/file/9t0gwc0unbbizn4/HP_8175A_HPJ_Apr_1987_Article_%2528pgs_4-20%2529.pdf/file
https://www.mediafire.com/file/6oe78ntawszrnmk/HP_8175A_%2528Opt_001%2529_Operating_and_Programming_Manual.pdf/file
https://www.mediafire.com/file/juu7yxhq6mz1oal/HP_8175A_%2528Opt_001%2529_Service_Manual.pdf/file
https://www.mediafire.com/file/mcpt3jlletrg46v/HP_8175A_%2528Opt_002%2529_Operating_%2526_Programming_Manual.pdf/file
https://www.mediafire.com/file/qh1ordpfsakvmht/HP_AN341-1.pdf/file
https://www.mediafire.com/file/ykiy1mnmpeswash/HP_AN341-2.pdf/file
garrettm:
For those thinking of picking one of these up, do note that there are two HW issues you will likely need to fix:

1) The NiCd back-up battery (2.4V, 110mAH, 2/3 AA) will be dead and possibly have leaked electrolyte onto the processor board. Two 1/3 AA or AAA NiMH batteries (connected in series) or a 5.5V 0.1 to 1 farad super capacitor can be used as a replacement. The super cap is likely to have less issue with leaking corrosive gasses/fluid in the event of a failure.

2) The power supply will need a replacement Y2 rated 22nF film capacitor (15.5mm pitch, 7mm max thickness). I would avoid using a RIFA/Kemet replacement part, seeing as how they still fail from cracking. Clearly, they don't care to fix the long standing issue with the faulty conformal coating.
garrettm:
Finally, the HP 8175A was intended to work with the HP 163x logic analyzers as a "stimulus response system" (see HP AN 341-1). My unit came with inverse assembler files for the 163x logic analyzers. I have archived the files I recovered off the floppy disks here:

https://www.eevblog.com/forum/testgear/searching-for-a-hp-1630-hp-1631-inverse-assembler-files/msg4039384/#msg4039384
wn1fju:
Thanks for the reminder about the batteries.  I haven't used my 8175A in years and when I opened it up to take a look, I found that I had already removed the battery on the uP board and ran two wires to a plastic battery holder that I mounted near the fans (I don't remember doing this).  When I opened up the battery cover, the two AAA NiMH batteries that I installed years ago had leaked.  But at least nothing in the 8175A got destroyed. 

What I really should do one day is get a bunch of little red stickers and put one on the front panel of each of my test equipment pieces that have any kind of batteries inside.  I simply can't keep track of which pieces do have them and when I ultimately come across one, the batteries are usually toast, including the ones that were used to replace the original.