I've just purchased an HP8590A Spectrum Analyser from that well known auction site for over £250.00 including the postage and I'm working my way through the installation and verification manual to check it out as it was listed as Used / Fully working and it completes the CAL:FREQ and CAL:AMPD and the CAL:YTO DELAY all OK
but when i do the CONF TEST it throws lots of errors eventually failing and Throwing out SRQ 110 - Hardware Broken
so my first question is can this unit be described as Fully working when it fails the CONF TEST routine with a Hardware broken message being returned?
and my second question is where do i start in restoring the unit to full working condition.
MostlyHarmless
Uunoctium:
Mine is an 8591E, but there should be no huge differences.
I have no access to manual at moment (stays on NAS in workshop) but if BW shape means IF filter curve is misaligend, you have to readjust. Instrument & manual guides you through that procedure - you only need a couple of jumpers. I used ordinary from pc-mobo.
After that, an complete Autocal-procedure is necessary. Should run approx 30 Minutes. But before that, check the Cal-Out for proper frequency and Amplitude. I used a counter with ext ref +Rb for the 1st and generator with trusted outputlevel + 2Ch Scope for 2nd by simply overlay the two traces.
If you decide to replace the backup battery on memory board, do'nt trust manual! There is something written, that the miracle supercap bypasses missing battery for a couple of days. This implies that it is working :'(
807:
Did you press the CAL STORE button after the CAL FREQ & AMPTD had finished?
While doing the CONF test, did you keep the CAL OUT connected to the INPUT connector? Did you see various signals on screen, similar to when doing the CAL test?
Why do you consider SRQ 110 to be "hardware broken"?
It cannot be described as "fully working" when it fails the CONF test. It's up to you if you want to try & realign it yourself, but I would still let the seller know. If you decide to keep it, then ask the seller for a partial refund.
I bought an 8591E around 15 years ago. I remember having to align it. But mine failed on just 1 parameter, not many like yours. I remember having to glue some jumpers to sticks in able to reach the boards. It was relatively simple, following the instructions in the service manual.
MostlyHarmless:
Thanks for your reply
I did indeed save the calibration after CAL:FREQ and CAL:AMPTD had finished.
I did have the CAL OUT feeding into the RF Input when I ran the CONFIDENCE test.
I consider SRQ 110 as "Hardware Broken" as that is what I found in the 8590A Portable RF Spectrum Analyzer Installation and Verification Manual, Appendix B, Table B.1
Thanks for the advice on contacting the seller, I had provisionally decided I would keep the unit and ask the seller for a partial refund but your advice has convinced me that is the best way forward
I've attached an image of my Cal Out on the display, I've checked the Frequency and its near enough at 299.893 and it shows the fundamental(@300Mhz) and four more harmonics up to the fifth(@1500Mhz), so i think my Cal Out is good enough for now
the first message i get out of the CONF test is POS-PK Fail, i guess i will start there using the Service manual and work my way through the Errors thrown out in the CONF test...
Ok, so consider this Stoopid question no.1 - what is the huge peak around 8 - 10 Mhz, is it the internal 10Mhz reference?
MostlyHarmless
ccktek:
I used an 8590a for several years. Initially it had confidence BW fails similar to yours, but the CAL procedures passed. I got it to pass the confidence self-test by carefully aligning the IF filter/amps. The factory alignment procedure didn’t always work quite correctly, so I had to fudge some of the adjustments so get the bandpass curves to look right. (I’m not sure about the other fails on your list.)
I found only one correct version of the service manual – it should be specifically for the 8590a, not the 8590 series. I believe this is correct:
https://ia800206.us.archive.org/32/items/hp_8590A_ServiceGuide/8590A_ServiceGuide.pdf
I agree with others that your 8590a can’t be considered to be fully working.
The big spike on the left represents DC and is normal when the start freq is set to 0.
Best of luck with your challenging project.