Re: FAN replacement
Out of curiosity I did an experiment to see if the damn stock fan is oversized in a configuration
without any internal oven.
As posted above I installed a NTC NTCASCWE3103F on the Q1 heatsink in the PSU which is the hottest guy in town.
I then wired the NTC terminals outside and started to measure the temperature in a closed case.
I chose the quietest fan 40x40x20 on Digikey in stock 2 wire, the
HA40201V4-1000U-A99 4500 rpm
Here the measurement results, Tamb was about 21C.
- unit plugged to main but power off

thermal equilibrium at 2150R --> Q1 Heatsink Temp 63.83C
- unit turned on and counting a 10MHz on CH1

thermal equilibrium at 1850R --> Q1 Heatsink Temp 68.18C
Based on that I keep that small fan and my hears are thanking me so much.