Hello
I made a mistake
The DE-5000 that I own measures too low the resistances of 10 ohms
I wanted to set VR2 which did not change anything, but when you change it, the value of the point TP1 changes
I got 1,300 Volts by positioning VR2 as originally (more or less)
Can give me the value you find on your devices
Thank you
A user who likes this product
-
Hi,
I measure 1.30688V at my DE-5000 with a Brymen 869S.
-
OK
Thank you for this information.
If I could have another answer, that would be reassuring.
If there is a French on the forum, I translated the whole manual into French.
Thanks again
-
1.30945V measured from TP1 to Battery Ground (on PCB) with a Fluke 8845A. After a few minutes to ensure settling. PCB temperature approximately 27C (warmer morning) in case that affects anything...
cheers,
george.
-
OK
Thank you for all those informations.
I will be able to set VR2 to this value.
If one of you knows what it is, it would be good for everyone.
For me, it's great, I think my DE-5000 will recover confidence
Thanks again to everyone
-
If there is a French on the forum, I translated the whole manual into French.
Thanks again
Hi there,
I just received my DE-5000 this morning.
Gege1244, if you agree to share your work, please post your translated manual here, it might help me and other frenchies. ;D
Kind regards
-
Check this out -> https://www.eevblog.com/forum/testgear/mastech-ms5308-lcr-meter-with-esr-measurement-on-discount-at-the-moment/msg1542980/#msg1542980 (https://www.eevblog.com/forum/testgear/mastech-ms5308-lcr-meter-with-esr-measurement-on-discount-at-the-moment/msg1542980/#msg1542980)
-
Hi there,
Thanks for the link.
Actually, I was just looking for a french version of the DE-5000 manual.
Regards.