My favourite 7904A failed
. Had it for about 10 years now and liked it very much. Replaced it with another 7904A from storage, but would really like to fix it.
Symptoms are: Scope starts, fan spins up, all voltages on PSU are present, but slightly low before PSU goes into ticking mode after about 2-3 seconds. It then ticks a few times, restarts and repeats the cycle at faster rate about once every 1.5-2 seconds.
(+/-50V rails => +/-47V, +/-15V rails => +/-14.7V, 5V rail => 4.7V, measured simultaneously with 5x HP-3456A with max/min recall in STAT mode)
Because of good voltages on main rails, I excluded a obvious short somewhere. Upon further investigation, I smelled electronic smell from HV compartment behind CRT cathode and noted that R14 (20Ohms current limit for HV inverter) gets rather hot.
Now, since most parts there are floating on -3000V, I'm a bit reluctant/scared to probe in it . Unfortunately, because of insane prices, I also don't have nice HV probes like Tek 6015. I could hack some simple HV divider from 10M resistors however.
Before I start to take it apart any more, is there any good strategy to further isolate faults? Some known bad parts in there? I somewhat suspect the anode HV multiplier to be bad (which would suck hard), it should be easy to disconnect it (1 wire) and I think the scope should run fine without it, in case it's bad(without screen of course). Any ideas?