I just got my LOM-510A calibrated by IET labs. The meter is 25 years old and had no signs of previous calibrations. Cal data shown below. The stability of this meter at 1 micro Ohm resolution is truly amazing. I have done enough testing with it to know that anything other than a rock steady reading or a perfectly consistent LSD bobble between two readings means something thermal is going on. This seems like the "Power Designs" of micro-Ohm meters, no frills, no programming, just exceptional accuracy even by today's standards.
I also made a stupid low $200.00 offer on another unit on EBay and got it
. Perfect condition and I checked/calibrated it against the IET calibrated unit as shown below. All ranges were in spec as recieved. Even though I adjusted the second unit to agree exactly with the calibrated unit on all ranges, some could argue there could be transferal errors. Even if I add 1LSD to the calibrated units error on each range it is still well within the .02% of reading + 2LSD + 2microOhm specification. So I feel confident the second unit is well within spec.
Calibration data
Both units warmed up and on the same range and are measuring in switched mode. Cables are clamped so that nothing moves when I switch the plug from one unit to the other.
Piece of tinned copper wire used for the .010 Ohm test. I just adjusted the spacing of my ultimate Kelvin probes to get a nominal 10 mOhm reading.
The Thermal coefficient of resistance of the copper wire requires the use of a insulated box to prevent variations from drafts and my body heat. just made one out of a piece semi rigid foam.
Chunk of iron holding the foam box tight to the table top.
The calibrated meter with the reading stabilized at 10.022 mOhm. The probe plugs are then interchanged between the two meters.
The second meter's 20 mOhm range trim pot is adjusted until the reading is stable at 10.022 mOhm. The probe plugs are interchanged a couple of times making sure the readings are identical between the meters. The meters cover is kept on all the time except to make a trim pot adjustment to keep the meter internal temperature consistent.
Comparison and adjustment on the other ranges was done using my decade resistance box. These are 25 ppm resistors and are heat sinked so they were stable without any additional thermal control.