| IFR Com 120A destortion meter error at boot-up |
| (1/1) |
| sfoweatherman:
Question: During my Com 120A boot-up/initialization, it reports an error on the screen regarding the distortion meter. If you go into the test procedures, and do the full test, the message on the destortion meter says: PASS. I've been told that it is a capacitor problem. If that is so, would the problem be on the controller tray? Thanks in advance, David |
| M0HZH:
Hello, Indeed, it is most likely an issue caused by a failing capacitor. The Baseband Tray handles the distorsion meter function (as well as other functions, which you might see fail soon) and it was plagued by a batch of bad capacitors. Just take the tray out and replace all of them, about 50 pcs. I also did the Digital, Analyzer and Generator trays on my IFR COM120B. Let me know if you need documentation, I have the manuals & schematics. Cheers, Razvan |
| sfoweatherman:
Thank you sir for the info. I will asses the board. Since the caps are all extra mini in size, I'll try to find the culprit...David |