| Products > Test Equipment |
| Inside the new Korad KA3305P linear PSU |
| << < (31/34) > >> |
| SantaClaw:
Yeah, well, I just did a voltage measurement whilst shorting it out.. :-DD It did 1.2v at 7 amps I don't think there is any form of current limitation on it. |
| stevelup:
The question really is what voltage yours outputs at 3A. Mine drops below 3V at just 1.9A load. Any chance you can apply a load instead of shorting it out? |
| SantaClaw:
I don't really have anything that can pull 3A, but I can tell you mine starts out low, at 4.918v |
| Scottjd:
--- Quote from: stevelup on November 30, 2019, 05:01:08 pm ---The question really is what voltage yours outputs at 3A. Mine drops below 3V at just 1.9A load. Any chance you can apply a load instead of shorting it out? --- End quote --- Hi Steve, I just ran a 3 amp load on mine. A short across a shunt in a multimeter will cause the voltage to drop and take all current over the rated 3A. Channel three doesn’t seem to have much in the over current protection. Without load activated the voltage reads 4.9965 from the sense wires. A constant current 3A load the voltage is 4.9464, I ran this for 4 minutes. A constant wattage at 15 watts will drop the voltage to 4.935 with the amperage at 3.039 A constant voltage load with the CV at 4.94 will draw 3.004A, total of 14.84 watts Basically, it looks like my channel three is stable at 14.80 watts, anything over and the voltage will dip below 4.9VDC If I turned the load up to 3.45A the voltage drops will cut off. Do you own a Koran unit? How old is your unit? Do you use channel 3 often? |
| GuidoK:
I did a little test with my electronic load on channel 3 where I started the load at 2,5A and incremented the load with 0,1A every 4sec. You can see that the current gets limited when it reaches 3,2A. (at 3,1A the voltage is still dead on 5V and at 3,2A the voltage is dropped at well below 0,5V) I have the rebranded RND unit from Distrelec |
| Navigation |
| Message Index |
| Next page |
| Previous page |