It is C46, 0.005uF = 5000pF 1.6kV in the schematic of the DX40 US version.
In my set, the cap was rated at 47pF 400V. That looks like grossly wrong rated cap, the previous owner put in.
Would it have contributed the serious over current fault in the set?
I did ask before but don't think I ever got an answer, what are you testing the capacitors with and what measurements are taking? If you are only using a meter that can read the capacity then that is not sufficient, yes it will show up any significant drift in value which may or may not be linked to your problem. You need to do a check on their ESR values using a meter capable of reading ESR and you also need to be able to check the capacitor for leakage and that involves testing with a far higher voltage then and ESR meter or a DMM meter is capable of supplying. You need to use something like a Heathkit C-3U bridge to test for leakage.
C46 could well be a smoking gun as with that amount of drift and voltage, it is more than likely to be leaking and could be the cause of your problem. Replace it and see. You could try, after discharging it, reading its resistance using a normal DMM and see if it is reading anything, if so replace it.
Ahh.. sorry I missed that question I think. I quickly drop by to the forum, and scanned msgs, so tend to miss some posts and questions at times.
Yeah, I do have an ESR reading meter which is BK-328, it displays ESR value. The fact is that the C46 is a new capacitor put in by the previous owner, but somehow he put the wrong valued one, because I can see the cap being very recent modern one, only replaced a few months ago?
Anyhow, I cannot tell now if it is directly linked to the problem of the over current without powering it on and testing it. I was looking into my junk box, but couldn't find any good matching value of the cap for the replacement. So I had to order some caps from eBay, and will take a few days to arrive, and I will wire the new cap on, and will power on for testing.
I am not sure if it is worth paying for the dedicated cap testing device like the vintage Heathkit. The problem is that they tend to be grossly over-priced for collectors value rather than what they do or how well they do it.
And it is quite impractical to desolder all the caps from the set, and go through the leakage testing when troubleshooting problems. You just try to identify the problematic areas, and pick out suspicious components and do testing, and if in doubt or problematic for sure, then replace them, that is my strategy. And capacitors are usually very cheap, so it is easy to replace them, if they look old or have usual suspicious tell tale signs for being bad.
And I am sure that reason that you are reading the capacitance is that when the value has drifted a lot from the nominal value, that would be the first sign that the cap is not good anymore, so it is vital to get the capacitance readings I thought.
All those old paper wax caps, when you take off from the set, and read the capacitance, they are usually well off from the nominal value, which tells a lot.
But really, this C46 was a very tricky cap, because it looked shiny new, and I took it for granted as a good cap, which turned out to be well out of value, not because it gone bad, but because it was simply a wrong cap for the junction in the circuit. If I have not tested it just purely by chance trying out my newly arrived DMM, looking at the schematic, it would have never been noticed forever.