| georgd:
I need an insulation tester that can be controlled from a computer, with 1000V test voltage. This would be part of an automated production line. Could someone recommend such a device? Thank you, Georg |
| Tjuurko:
UT512 can connect to a computer. https://eleshop.eu/uni-t-ut512-insulation-resistance-tester.html I don't know about the possibility of control from a computer. |
| jonpaul:
"insulation tester" is a wide term. Safety? AC/DC 1000V? leakage current to earth? range of current? application eg transformer/motor insulation test? Medical equipment compliance? Instrument manufacture? For old line USA firms, see Rod-L, Hipotronics, Stangenes Industries... Jon |
| georgd:
Here is a more detailed description of requirements: - The factory produces mains (230/380 V) cables equipped with cable head on both side - Instrument is intended to use in a factory line as test gear. - Test test voltage is 1000V DC and the resistance between the cable conductors needs to be measured. Georg |
| jonpaul:
https://www.megger.com/en/type/electrical-insulation-testing ROD-L Danbridge HiPotTronics https://www.hubbell.com/hipotronics/en/products/electrical-electronic/test-equipment/instruments/cl/547141 enjoy, Jon |