Products > Test Equipment

Interfacing the HP3490A with AR488

(1/2) > >>

Joel_S:
Hello,
I am trying to control an HP3490A DVM from a PC via GPIB with an Arduino-based AR488, using PuTTY as the user interface.
So far, two things already work :
- When the 3490A is in "talk only" mode, measurement data is received and displayed as expected
- When in "addressed" mode, I am able to control the functions of the 3490A (select functions, switch ranges etc ...) as explained in table 3-7, page 3-10 of the 3490A manual.
However, when in "addressed" mode, I cannot receive measurement results over the HP-IB. It seems that the 3490A requires the ATN line to be activated, but the AR488 manual doesn't provide any information on how to do that.
Any help would be appreciated ... the 3490A is a very nice beast !
Thanks,
Joel Setton

Gertjan:
Did you ask for the answer with ++read?

First you send a command to the DVM to request a measurement, the DVM will take the measurement, and put it in it's buffer. When you send as next command ++read, the buffer will be read, and the answer sent to your terminal progam.

Alternatively you can start with sending ++auto 1 to the AR488. With this configuration command, the AR488 will automatically try to read the DVM buffer after a measurement command is sent.
Read the AR488 manual for the details.

regards & good luck, Gertjan.

Joel_S:
Hello Gertjan, and thank you for taking the time to respond !

I have done quite a bit of experimenting along the line of your suggestions. I first connected the AR488 to an HP 3457A multimeter, and followed this sequence:
++ mode 1
++addr 12    // this is the HP-IB address of the 3457A
++read
and I receive the reading from the 3457A:
1.0000034E+00

I also used the ++auto 1 command, as you suggested. It worked ! Readings are displayed automatically after each command, for example typing ?[CR] immediately returns the voltage reading from the 3457A.

Unfortunately, this does not work with the 3490A. I can "talk" to it to select various unctions (DCV/ACV/Ohms) and ranges, but I can't get the 3490A to talk back. It is an old instrument, and I understand its handling of the HP-IB is not the same as more recent models such as the 3457A. I'm still trying !

Thanks again,

Joel

ch_scr:
Have you checked you didn't by accident set the unit to "listener" only?

Joel_S:
Hi !
I didn't know the instrument could be set to "listener only".
There is a switch on the 3490A rear panel, to select between "Addressable" and "Talk Only".
I set it to "addressable".
When it was set to "talk only", the voltage measurements were received OK by the AR488 and displayed by PuTTY, but of course I could not control the instrument.
When set to "addressable", I can control the instrument (selecting DCV / ACV / Ohms, ranges, etc...) but measurement results are not received by the AR488.

Joel

Navigation

[0] Message Index

[#] Next page

There was an error while thanking
Thanking...
Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod