Author Topic: Agilent E4400B Reverse Overload  (Read 1012 times)

0 Members and 1 Guest are viewing this topic.

Offline RCHRDMTopic starter

  • Regular Contributor
  • *
  • Posts: 125
  • Country: us
Agilent E4400B Reverse Overload
« on: April 22, 2019, 12:02:31 am »
I am embarrassed to say that I accidentally sent RF into the signal generator at about 50 watts.  This is higher than the overload circuit can protect, I think.

The resistance across the output N-connector is zero or open.  It should be 50 Ohms, right (with power off)?

The signal generator still has a normal output as far as I can tell.

Does anyone know how to check this out and repair it?

I am looking through the service manual.

Richard
 

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 27165
  • Country: nl
    • NCT Developments
Re: Agilent E4400B Reverse Overload
« Reply #1 on: April 22, 2019, 12:32:29 am »
But it is still working? If yes, consider yourself lucky and carry on.
I have a similar generator; AFAIK that has a relay inside which disconnects the output in case of reverse power (or overload).
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline RCHRDMTopic starter

  • Regular Contributor
  • *
  • Posts: 125
  • Country: us
Re: Agilent E4400B Reverse Overload
« Reply #2 on: April 22, 2019, 01:34:21 am »
Thanks!  That's encouraging.

But shouldn't the RF output connector measure 50 Ohms (power off)?

Richard
 

Offline TheSteve

  • Supporter
  • ****
  • Posts: 3756
  • Country: ca
  • Living the Dream
Re: Agilent E4400B Reverse Overload
« Reply #3 on: April 22, 2019, 01:38:55 am »
I believe the output is AC coupled so you'll likely measure an open.
VE7FM
 

Offline RCHRDMTopic starter

  • Regular Contributor
  • *
  • Posts: 125
  • Country: us
Re: Agilent E4400B Reverse Overload
« Reply #4 on: April 22, 2019, 01:48:30 am »
SIGH OF RELIEF!

I tried testing the output with my oscilloscope with an output of 10 MHz and 10 mV.  Everything looks good!

I really appreciate your help.  That makes a bad evening turn into a good evening!

I will be more careful...

Ricahrd
 

Offline bob91343

  • Super Contributor
  • ***
  • Posts: 2675
  • Country: us
Re: Agilent E4400B Reverse Overload
« Reply #5 on: April 22, 2019, 04:04:40 am »
I recently did the same thing with my HP 8657B.  After resetting the reverse protector, the output was down about 20 dB although everything else worked.

I tracked it down to bad SPDT switch(es) in the frequency doubler.  Replaced those and now all is well again.  And learned a lesson on repairing SMT.
 

Offline RCHRDMTopic starter

  • Regular Contributor
  • *
  • Posts: 125
  • Country: us
Re: Agilent E4400B Reverse Overload
« Reply #6 on: April 22, 2019, 04:09:36 am »
Good work!  I have not worked on SMD stuff yet...