Author Topic: Sencore LC102 lcr meter  (Read 3110 times)
Offline bubbatechTopic starter
« on: March 05, 2018, 07:26:47 pm »
I bought a Sencore lc102 meter at a hamfest for an insane good price.  I calibrated it and it worked flawlessly for month.
Now, it says "error 7" when one attempts to zero it, or measure capacitance below 2uF (but not above). Inductance and ESR measurements are fine. The onset of the problem was sudden as if something folded. I have been working on this thing for about a month now, trying to diagnose the problem and it is absolutely kicking my ass. I have replaced the relays, the lm319 comparator, and for awhile suspected IC12, a flip-flop that addresses the current sources for capacitance measurements since only measurements below 2uf were affected (therefore, I suspect, screwing the ability to zero compensate). One theory was a power supply issue, but all the rails are nominal, except for the 12 volt unregulated rail which is at 13.8 volts, but I think that is normal. My most recent theory has been that there is a problem with one line in the control buss, but swapping the IC and checking all the components on that buss (F5) yielded nothing.  :-//
In fact, though, I am still guessing. I still don't even know which section of the circuit is problematic.
If someone who has some familiarity with this instrument could throw me a bone, it would be very grateful. I am running out of ideas.
This is such a nice instrument. I refuse to give up.  |O
Offline Johnny10
Re: Sencore LC102 lcr meter
« Reply #1 on: March 05, 2018, 08:23:37 pm »
Yes I had 13 something on mine.
I have to dig up my notes.
I had the most problems with those "split the wire" connectors.
And you have cleaned and measured the resistance through the Combination fuse/BNC input ?
My favorite LC meter!
« Last Edit: March 05, 2018, 08:27:45 pm by Johnny10 »
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
« Reply #2 on: March 06, 2018, 12:39:25 am »
Thanks for your reply. Yes, I checked the usual culprits. The input connections are clean and, prior to the failure, an error 4 indicating resistance out of spec did not occur. I have the full, original schematics for it, which are quite nice, but they do not specify nominal voltages and there is no discussion of the theory of operation, so it is not clear if some of the measured voltages are what they should be, since I never measured them before the unit failed.
Offline Johnny10
« Reply #3 on: March 06, 2018, 11:33:27 am »
Tell me which measurements you are looking for and I will check on mine.
I have had a few of these units and repaired two.
Repairing the display on one of these was my first foray into electronics repair 4 years ago.
Offline Johnny10
« Reply #4 on: March 06, 2018, 04:31:06 pm »
Looking over the specs of the LC102 pg 7.
There is range switching of Capacitor ESR :
Your problem must be related to that circuit.
The patent is available 4795966
"the present invention a new and improved method for measuring the equivalent series resistance of a capacitor by charging the capacitor from a constant current source"
So problem must be in the ESR circuitry.
Schematic on left hand bottom of 2000 Board
« Last Edit: March 06, 2018, 06:06:34 pm by Johnny10 »
Offline Johnny10
« Reply #5 on: March 06, 2018, 06:24:40 pm »
Looking at the Error Code 7 description.
The error code describes a 1uF limit to ESR.
« Reply #6 on: March 07, 2018, 03:35:19 am »
Thank you very much for your time. I have checked esr  measurements by connecting resistors across the leads and it appears to read ok, so that idea is not intuitively obvious to me (which could be why I have so far failed!) However, if the ESR section is involved in capacitance measurements, then it becomes much more plausible. I have looked at ic21 some, but nothing else. Tomorrow evening, I'll try to make some measurements, paying attention to CR1 and some of the lines leading from IC21 to those op-amps to see if any of the capacitors have shorted to ground.
Thank you!
« Reply #7 on: March 13, 2018, 09:11:09 pm »
It has been awhile since I worked on the instrument because I have been out of town. I am getting back to it, but there i one critical question. There is a 12V unregulated power rail. On that rail, I find 13.8V. If this is abnormal, it would explain why the current sources for low capacitance measurements never turn on because the base would always remain relative negative. On the other hand, 12V rails in battery powered devices can be as high as 13.8V normally. Do you see this in your working device?
Thanks!
Offline Johnny10
« Reply #8 on: March 14, 2018, 02:00:02 am »
Where are you measuring the 12v unregulated line?
I just opened my unit and have it sitting on bench.
« Last Edit: March 14, 2018, 02:46:38 am by Johnny10 »
« Reply #9 on: March 14, 2018, 02:03:22 pm »
Damn. I wish I had seen this sooner. Sorry.
You could measure it on pin 6 of P4 on the power supply board, or Pin 5 of P6 on the main board. These are connected to the output of the unregulated 12 V supply. It is plausible that this would be higher than 12V because I think it is used to charge the battery, but I need to make sure so I can rule it out.
In any event, I think the primary problem is that the line that should enable the relevant current sources, which are the cathode side of CR6 and CR7 on the main board should go logical low for  at least a short period (I think) when capacitance measurements are made or the instrument is zeroed. When that line goes more negative, it should drive current across R53 and R54, pulling the base of Tr11 more negative, turning it on. I have never seen these lines go low - ever. They stay rock solid at 4.8V. It is my feeling that this can't be right. If it is convenient, can you measure these lines when the capacitance button is pressed (or the zero switch) and tell me if they go low as well? That would be extremely helpful. If they never go low on your instrument, then I am chasing the proverbial wild goose.
Thanks!
Offline Johnny10
« Reply #10 on: March 14, 2018, 02:46:33 pm »
No Problems I cleared the bench for another project.
Pin 6 P4  on 3000 Board
14.77 Volts
