[SOLVED] NI MAX does not see my Siglent SPD3303
« on: August 25, 2022, 05:44:54 pm »
Hi,
I'm unable to connect to the PSU using NI MAX. Other Siglent devices (DMM, scope, AWG) are showing... Not the PSU. Entering its IP by hand has no effect.
I can control the PSU using EasyPower. I can ping it.
But no luck with NI MAX.

DHCP + static lease so the IP is always the same.

unlocked SPD3303X-E (=SPD3303X)
hardware : v3.0
firmware : SPD3303X V1.01.01.02.07R2

What am I doing wrong ?
« Last Edit: August 25, 2022, 10:18:45 pm by mushroom »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29335
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: NI MAX does not see my Siglent SPD3303
« Reply #1 on: August 25, 2022, 08:36:05 pm »
If rolled back to the X-E version does the problem still remain ?
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline mushroomTopic starter

  • Regular Contributor
  • *
  • Posts: 123
  • Country: fr
Re: NI MAX does not see my Siglent SPD3303
« Reply #2 on: August 25, 2022, 10:03:07 pm »
I reverted the PSU to X-E, re-upgraded to 1.01.01.0.2.07R2 (X-E firmware version) : it does not make any difference. Tested from two computers.
I also tried with another IP I defined on the device rather than assigned by the DHCP. Still no PSU on the NI utility even when trying to define it manually : "Could not open a VISA session to 'TCPIP0:nnnnnn blah blah'"
The DHCP (a PiHole) works properly. When validating the DHCP option on the PSU, I can see the IP changing for the static lease I defined in the DHCP.
I can telnet using port 4025 (seen here : https://www.eevblog.com/forum/testgear/spd3303x-e-scpi-interface-issues/), I get a beep when connecting, but have no echo so I don't see what I type.

[EDIT] problem solved !
I wasn't properly entering the parameters. In the dialog box, do not enter information in the "LAN Device Name". I was entering the Hostname in this field !
So, entering the IP or the hostname in the first field, and letting the second one alone is the right thing to do. RTFDI ! Read The F* Dialog Instructions !

https://www.siglenteu.com/application-note/verification-of-a-working-remote-communications-connection-using-ni-max/
« Last Edit: August 25, 2022, 10:22:02 pm by mushroom »
 