Products > Test Equipment

is this a fluke? (117 ghost? current)

(1/2) > >>

Anthocyanina:
I recently got a fluke 117 and I haven't used it much, and when i have, it has only been for circuits using 5v and for continuity, and the day it arrived, i also shorted some AA batteries with it to see the current go up, everything seemed fine afterwards and it was left turned off for a while. today i needed to measure some currents in the order of a few milliamps, and grab it for that, but when turning it on, i encountered this strange thing, it shows 4 mA DC with nothing plugged into any terminal, and when switching to AC current, while it displays 0.000, the first segment of the bargraph flickers on and off.

I opened it and took out the fuse, and it still behaved the same way. Then i changed the battery, and it showed 0.000 A on DC but the bargraph still flicked on AC, and after a few minutes, the 4mA started to show up again when on A DC.

Any idea what might be going on?
Thank you!

evava:
I think you should plug in the test leads, then look at display in DC mA range ?

(split input terminals)

Anthocyanina:

--- Quote from: evava on November 25, 2022, 09:57:56 am ---I think you should plug in the test leads, then look at display in DC mA range ?

(split input terminals)

--- End quote ---

Thanks, I've already tried shorting the terminals with the leads. This meter doesn't have a mA range unfortunately

skander36:
Maybe need calibration? (ofsetting)
My fluke 87V also show 2 mA on iddle and I have others that fluctuates on iddle (plus or minus 1-2 mA).

Anthocyanina:

--- Quote from: skander36 on November 25, 2022, 10:28:41 am ---Maybe need calibration? (ofsetting)
My fluke 87V also show 2 mA on iddle and I have others that fluctuates on iddle (plus or minus 1-2 mA).

--- End quote ---

Thank you, i tested various currents (through a 100 ohm resistor), and it does seem to be a constant offset! I do wonder what caused this offset to appear while it was just sitting there unused for a few days, hmm!

Navigation

[0] Message Index

[#] Next page

There was an error while thanking
Thanking...
Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod