Why do some multimeters have dedicated mV ranges?
« on: May 18, 2013, 05:55:51 pm »
I was looking at a Fluke 73iii I own and started thinking, why does it have a dedicated 300mV range?  I can understand why someone might want to measure a low voltage.  I'm not questioning the desire to measure sub 300mV ranges but really, why can't/isn't this incorporated into the autoranging DC voltage range?  Is there a significant change in input impedance in the mV range?
Incidentally I have seen this on several of my non-Fluke meters as well. 
 

Offline Monkeh

  • Super Contributor
  • ***
  • Posts: 7992
  • Country: gb
Re: Why do some multimeters have dedicated mV ranges?
« Reply #1 on: May 18, 2013, 06:11:58 pm »
Is there a significant change in input impedance in the mV range?

Only about 2.99GOhm, typically.
« Last Edit: May 18, 2013, 06:13:52 pm by Monkeh »
 

Offline RobomedsTopic starter

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: us
Re: Why do some multimeters have dedicated mV ranges?
« Reply #2 on: May 18, 2013, 06:36:19 pm »
gigaohms.... hmmmm....  :-DD
 

Offline Monkeh

  • Super Contributor
  • ***
  • Posts: 7992
  • Country: gb
Re: Why do some multimeters have dedicated mV ranges?
« Reply #3 on: May 18, 2013, 06:40:05 pm »
Your Fluke seems to be 10M on mV, though. Pays to check the specs..

Not sure why it has a separate range for the same impedance.
 

Offline RobomedsTopic starter

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: us
Re: Why do some multimeters have dedicated mV ranges?
« Reply #4 on: May 18, 2013, 06:50:34 pm »
Well the 87-iii and 73-iii show the same impedance in both the mV and V settings according to Fluke.  However, I do see that if I start the meter in high impedance mode the 400mV range does gain 3.99 Gohms of impedance.  By default the 87-iii has only 10Mohm resistance in either DC volts range.  That could explain why the 87 has such a range but not the 73-iii which doesn't have a high impedance setting. 


 

Offline ModemHead

  • Frequent Contributor
  • **
  • Posts: 731
  • Country: us
  • No user-serviceable parts inside.
    • Mr. ModemHead
Re: Why do some multimeters have dedicated mV ranges?
« Reply #5 on: May 18, 2013, 06:52:17 pm »
On Fluke auto-ranging DMMs, there is usually a 10Meg resistor in series with the input, and one of several other grounded precision resistors is switched in by the microprocessor to form a divider for 1:10, 1:100, etc.  But for the millivolt range, the 10Meg resistor is grounded, and the ADC input is taken straight off the input jack (1:1) through the ohms sense resistor or through a separate resistor of about 1Meg.  The input impedance thus remains at about 10Meg for all ranges.

Given this input divider reconfiguration, I suppose it's easier to just have another switch position.
 