Author Topic: Keithley 2000 TRMS converter noise  (Read 21902 times)

0 Members and 1 Guest are viewing this topic.

Offline syncTopic starter

  • Frequent Contributor
  • **
  • Posts: 799
  • Country: de
Keithley 2000 TRMS converter noise
« on: February 21, 2014, 10:56:56 pm »
Hi

I repaired a dead Keithley 2000. The power supply caps leaked and eaten PCB traces. It's working now again. :)

I would like to know the typical TRMS converter noise. It would be nice if someone with a Keithley 2000/2015 could measure it. Set the meter to ACV, 5.5 digits and short the input. The reading is the TRMS converter noise. I get 121 counts. The manual states 100 counts typically.
 

Offline ratsept

  • Newbie
  • Posts: 1
Re: Keithley 2000 TRMS converter noise
« Reply #1 on: February 21, 2014, 11:31:57 pm »
Hi,

I just tried my 2015 THD and I'm getting 127 counts. Hope it helps.
 

Offline syncTopic starter

  • Frequent Contributor
  • **
  • Posts: 799
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #2 on: February 22, 2014, 01:40:16 pm »
Yes, it helped. The TRMS noise of my unit didn't degraded.
Thanks!
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Keithley 2000 TRMS converter noise
« Reply #3 on: February 22, 2014, 03:10:12 pm »
5.5digit , 1 NPLC, REL + Auto filter
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #4 on: February 22, 2014, 03:33:41 pm »
I would like to know the typical TRMS converter noise. It would be nice if someone with a Keithley 2000/2015 could measure it. Set the meter to ACV, 5.5 digits and short the input. The reading is the TRMS converter noise. I get 121 counts.
I have never really checked this on my Keithley.
Therefore out of interest I just ckecked my 2015 and it is around 0.13mV and the Tek DMM 4050 is 0.04mV.
Is this an indication that the 4050 is 3x better in this respect?

The manual states 100 counts typically.
@sync, where in the Manual can I find this "100 Count" information?

5.5digit , 1 NPLC, REL + Auto filter
@TiN, I would be interested to see your values without "REL", if possible.

thanks
quarks
« Last Edit: February 22, 2014, 04:03:47 pm by quarks »
 

Offline syncTopic starter

  • Frequent Contributor
  • **
  • Posts: 799
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #5 on: February 22, 2014, 04:03:02 pm »
Keithley 2000 user manual, Page 2-19:
Quote
AC voltage offset
The Model 2000, at 51?2 digits resolution, will typically display 100 counts of offset on AC volts with the input shorted. This offset is caused by the offset of the TRMS converter. This offset will not affect reading accuracy and should not be zeroed out using the REL feature. The following equation expresses how this offset (VOFFSET) is added to the signal input (VIN):

Displayed reading = sqrt(VIN² + VOFFSET²)

Regarding if the 4050 is 3x better. I don't know. Maybe they use a less noisy analog TRMS converter or a different TRMS conversion method.

@TiN: Many thanks for hosting the K2000 schematic. It helped a lot during repair.
 

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #6 on: February 22, 2014, 04:12:18 pm »
Keithley 2000 user manual, Page 2-19:

thanks sync
it is the same in the 2015 Manual (see att.)
 

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #7 on: February 22, 2014, 04:24:53 pm »
because I made my first test in 100mV range, I now repeated it in 1V range
Keithley 2015 is now 1.27mV
TEK 4050 is 0,33mV
« Last Edit: February 22, 2014, 05:23:01 pm by quarks »
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Keithley 2000 TRMS converter noise
« Reply #8 on: February 22, 2014, 04:32:32 pm »
Time to open KI fanclub? :) Btw, this is after AC CAL (internal AC calibration procedure)

KI2002:

200mVAC rangeInputs shorted0.052xxx mVAC RMS10 NPLC, 8.5d, no REL
200mVAC rangeInputs open6.559xxx mVAC RMS10 NPLC, 8.5d, no REL
2 VAC rangeInputs shorted0.000142xx mVAC RMS10 NPLC, 8.5d, no REL
2 VAC rangeInputs open0.00657xxx mVAC RMS10 NPLC, 8.5d, no REL
20 VAC rangeInputs shorted00.0032xxx mVAC RMS10 NPLC, 8.5d, no REL
20 VAC rangeInputs open00.0084xxx mVAC RMS10 NPLC, 8.5d, no REL
200 VAC rangeInputs shorted000.010xxx mVAC RMS10 NPLC, 8.5d, no REL
200 VAC rangeInputs open000.0146xx mVAC RMS10 NPLC, 8.5d, no REL
750 VAC rangeInputs shorted0000.048xx mVAC RMS10 NPLC, 8.5d, no REL
750 VAC rangeInputs open0000.051xx mVAC RMS10 NPLC, 8.5d, no REL

Sorry, 2001 off now, messing with NVRAMs in it ...
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline robrenz

  • Super Contributor
  • ***
  • Posts: 3035
  • Country: us
  • Real Machinist, Wannabe EE
Re: Keithley 2000 TRMS converter noise
« Reply #9 on: February 22, 2014, 04:32:49 pm »
What I found in this old thread implies the 8846A AC performance may be better than average

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #10 on: February 22, 2014, 04:50:41 pm »
@TiN, thanks for the numbers.
but can you please recheck beginnen from 2V range your values are probably VAC and not mVAC

@robrenz, yes once again the Fluke8846A/TEK DMM4050 show amazing good results. 

Just to see what my 8.5 Digits meters show with this quick test, I fired them up.
Whatever this really means, the little 4050 is second best!!!

 
« Last Edit: February 22, 2014, 05:32:35 pm by quarks »
 

Offline casinada

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: us
Re: Keithley 2000 TRMS converter noise
« Reply #11 on: February 22, 2014, 05:48:28 pm »
On my 2015 it shows:
100mV range: 000.0084 mVAC
1V       range: 0.00074 VAC
10V     range: 00.0073 VAC
100V   range: 000.073 VAC
750V   range: 0000.71 VAC
 
 

Offline syncTopic starter

  • Frequent Contributor
  • **
  • Posts: 799
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #12 on: February 22, 2014, 05:57:03 pm »
On my 2015 it shows:
100mV range: 000.0084 mVAC
1V       range: 0.00074 VAC
10V     range: 00.0073 VAC
100V   range: 000.073 VAC
750V   range: 0000.71 VAC
Interesting. That is considerably less than the 120-130 counts we had so far. Is this a newer unit?
 

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #13 on: February 22, 2014, 06:03:26 pm »
On my 2015 it shows:
100mV range: 000.0084 mVAC
1V       range: 0.00074 VAC
10V     range: 00.0073 VAC
100V   range: 000.073 VAC
750V   range: 0000.71 VAC

That is interesting and is well below the 100 number from the manual
(only the 000.0084 mVAC should probably be 000.084 mVAC).

I wonder if the test results can be improved somehow.
Have you observed if the "Offset" changed over time?

I hope some more members do this test, and Report their values.
« Last Edit: February 22, 2014, 06:35:33 pm by quarks »
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Keithley 2000 TRMS converter noise
« Reply #14 on: February 22, 2014, 06:09:13 pm »
I hope some more members do this test, and Report their values.

K2010, eta: Monday afternoon.

Offline mimmus78

  • Supporter
  • ****
  • Posts: 676
  • Country: it
Re: Keithley 2000 TRMS converter noise
« Reply #15 on: February 23, 2014, 12:19:31 am »
Mine numbers:

100mv 000,032
1V       0,00028
10V     00,0028
100V   000,027
750V   0000,26

so I'm the winner of K2000 contest?
 

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #16 on: February 23, 2014, 08:14:51 am »
Mine numbers:

100mv 000,032
1V       0,00028
10V     00,0028
100V   000,027
750V   0000,26

so I'm the winner of K2000 contest?
That looks really good.
Can you give us more details of your meter and setup?

Here are links to some more info about this topic:

https://www.eevblog.com/forum/testgear/dmm-measuring-ac-volts/msg233469/#msg233469

http://www.ni.com/white-paper/2966/en/#toc6
« Last Edit: February 23, 2014, 11:15:13 am by quarks »
 

Offline robrenz

  • Super Contributor
  • ***
  • Posts: 3035
  • Country: us
  • Real Machinist, Wannabe EE
Re: Keithley 2000 TRMS converter noise
« Reply #17 on: February 23, 2014, 03:23:34 pm »
Numbers from my trusty 8846A :-*  (yes, I kiss my meter)

Shorted inputs, not zeroed out, high resolution, stat mode of 10 readings.  No typo's possible with screen shots ;D

100mV


1V


10V


100V


1000V
« Last Edit: February 23, 2014, 03:29:29 pm by robrenz »
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Keithley 2000 TRMS converter noise
« Reply #18 on: February 23, 2014, 04:43:09 pm »
robrenz's weaker brother, the DMM4020, not zeroed, slow rate:

Code: [Select]
200mV 0.000mV
2V 0.00024V
20V 0.0018V
200V 0.011V
750V 0.04V

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #19 on: February 23, 2014, 04:58:06 pm »
Numbers from my trusty 8846A :-*  (yes, I kiss my meter)

Shorted inputs, not zeroed out, high resolution, stat mode of 10 readings.  No typo's possible with screen shots ;D

wow, that is 3x lower than my own measurements :-+
 

Offline KJDS

  • Super Contributor
  • ***
  • Posts: 2442
  • Country: gb
    • my website holding page
Re: Keithley 2000 TRMS converter noise
« Reply #20 on: February 23, 2014, 05:17:29 pm »
I just measured three 2015THD on the 1V range

measured counts were 171, 80, 71


Offline syncTopic starter

  • Frequent Contributor
  • **
  • Posts: 799
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #21 on: February 23, 2014, 06:31:34 pm »
Robrenz, these numbers are very low. Do you know how good it measures at the low end of the scale? My 34401A shows also very low noise with shorted input. Less that 6ppm of the range. But the measurements drop off dramatically under 0.4% of the range, 0.5% on the 100mV range. IMHO these low noise readings are faked on the 34401A. Geller Labs have observed this too (http://www.gellerlabs.com/34401A%20AC%20zero.htm).
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Keithley 2000 TRMS converter noise
« Reply #22 on: February 23, 2014, 06:42:17 pm »
Code: [Select]
                          HMC8012, ADC=Slow, AC Filter=Med, Not zeroed
Range Mean         Min         Max         Std Dev         Pk to Pk # Measures (1000=~3min20sec)
400mV 0.062mV 0.058mV 0.065mV 0.001mV 0.007mV 1000
4V 0.00072V 0.00070V 0.00073V 0.00001V 0.00004V 1000
40V 0.0080V 0.0078V 0.0081V 0.0000V 0.0003V 1000
400V 0.080V  0.077V  0.082V  0.001V  0.005V  1000
750V 0.06V    0.05V    0.06V    0.00V    0.01V    1000




Offline robrenz

  • Super Contributor
  • ***
  • Posts: 3035
  • Country: us
  • Real Machinist, Wannabe EE
Re: Keithley 2000 TRMS converter noise
« Reply #23 on: February 23, 2014, 08:57:06 pm »
Robrenz, these numbers are very low. Do you know how good it measures at the low end of the scale? My 34401A shows also very low noise with shorted input. Less that 6ppm of the range. But the measurements drop off dramatically under 0.4% of the range, 0.5% on the 100mV range. IMHO these low noise readings are faked on the 34401A. Geller Labs have observed this too (http://www.gellerlabs.com/34401A%20AC%20zero.htm).

I addressed this in the link I posted on reply #9 of this thread.  Its here again for convienience

Offline syncTopic starter

  • Frequent Contributor
  • **
  • Posts: 799
  • Country: de
Re: Keithley 2000 TRMS converter noise
« Reply #24 on: February 23, 2014, 09:46:38 pm »
I addressed this in the link I posted on reply #9 of this thread.  Its here again for convienience
Sorry, I missed that. Wow, the 8846A is impressive! I subtracted the 12.5uV noise from your measurements.
VMEAS = sqrt(VREADING² - VNOISE²).

  VIN     VMEAS
 20uV    19.1uV
 30uV    29.7uV
 40uV    40.7uV
 50uV    50.7uV
 60uV    61.3uV
 70uV    74.5uV
 80uV    82.3uV
 90uV    92.6uV
100uV   102.9uV


 