chickenHeadKnob:
Is it really the same analogue multi-slope AD board (or section) in all of 2001, 2002. 2010 ? This would imply possible home uprating on 2001's hmm....

Noticed in one of TiN's reveals they were using OP177's, not a bad op amp but I thought there might be better now-a-days.

Thank you for the pics
dr.diesel:
Update, finally back (received March 28th) from Keithley!

This is my first cal process outside of the corporate environment.  It took two rounds with Keithley to even get a cal quote, first one came back as a quote for a new 2010!  Anyhow, here is the breakdown:

Calibration std price:  $315, Optional Data: $65, $180 for a firmware update and $12.50  return shipping.  I had no personal reason for the data, just thought it might be of some interest to others here.  From the documentation I knew there was at min 6 newer firmware revisions, so I opted for the latest.

Keithley received the meter on March 10th.  On March 18th I received an email stating the meter was "noisy" and required repair with a new quote:

Repair with Cal: $1120, Optional Data: $65 and $12.50 return shipping.  It would have cost me the $180 and $12.50 to have them simply return it without repair.  Note this is total, not in addition to the above first cal quote.

For those interested in what a "noisy" Keithley 2010 with a 14 year old cal reads:



I tested the meter thoroughly before sending in for cal, but I'm not going to call them out, I am for the most part satisfied with how to went.  The repair included an entirely new board, rubber keyboard membrane, and I'm pretty sure VFD.  A lot of good those new caps did me, ha ha.

Summary:

Used 2010 for: $1000
New PS Caps: ~$30
Repair/Cal total: $1197.50

In total about 2/3s of a new unit.





Vgkid:
Thanks for the update, it is interesting to note the ranges that it failled.
TiN:
Thanks for data, interesting to find lots of values in 8.5 digit mode. None of my 2001's data cal have even 7.5 there.

180$ for replacing ROM chip with firmware? Wow :)
I had mine 2001 replaced for free, no cost increase.

Do you know what was repaired?
From cost it would look like they replaced whole board maybe.
dr.diesel:

--- Quote from: TiN on March 29, 2014, 04:58:23 pm ---Do you know what was repaired?
From cost it would look like they replaced whole board maybe.

--- End quote ---

Yup, the 2010 is a single board aside from the VFD.  The board was for-sure replaced and I think they also swapped the VFD, noticeably brighter and no more soft spots around the edges of some of the characters.