Keithley 2231A-30-3 Communication
(1/2) > >>
simba15:
Hi there,

I am working with a large number of the Keithley 2231A-30-3.

I would like to communicate with it VIA PC but Tek, states you must use the Keithley USB adapter ($$$)

I have found the pin-out of the propriety connector (2=rx,3=tX).

Has any one been able to get this power supply to communicate to their pc? ( may only respond to read only commands)

Thanks.
simba15:
So I connected a FTDI TTL adapter tx/rx and tried with EEZ studio to get a response for a *IDN? command.

I cant get a response in the software but i can see that both Tx/Rx lights illuminate. (there is a response of some kind.)

Any ideas where the issue could be?

I will try with "Test controller" software.
alm:
Did you try different baud rates / flow control / parity / stop bit configurations? If it outputs anything on power on, you could capture that with a scope or logic analyzer and check the repetition rate and count the number of bits.
simba15:

--- Quote from: alm on January 11, 2022, 02:35:30 pm ---Did you try different baud rates / flow control / parity / stop bit configurations? If it outputs anything on power on, you could capture that with a scope or logic analyzer and check the repetition rate and count the number of bits.

--- End quote ---

So the Manual Lists the Baud rate and flow settings 9600,8,N,1 ( this is what the program is set for)

If i use a incorrect baud rate 57600 or other, the RX light dose not illuminate.

I will try and scope the RX Line.
BennoG:
was you been able to communicate with the device ?
It uses 3.3V level RS232 signals at 9600 baud.

I found it to be very slow for the measurements of the output Voltage and Current (300+ ms per query)
Have you come around this problem ?

Benno