Author Topic: keithley 2400 C power issue  (Read 1818 times)
Offline richcj10
« on: April 24, 2013, 03:30:45 am »
I have acquired this piece B/C it wasn't working. (nothing on the screen, no beep) I have found out that the AC in to DC (12v & 5v) is working but the 12V DC to AC to +-15, +-30, etc isn't.
I think the problem is at the front end of the transformer. does anyone have the schematic?
Thanks,
Rick
Offline MetraCollector
  • Country: cz
Re: keithley 2400 C power issue
« Reply #1 on: April 24, 2013, 10:12:30 am »
As I know, schematics are not available, I was trying to find some for K2000 with no success.
Anyway, service manual is available through G. and there are some flow chart for repairing these Keithley instruments.
GL !
Offline free_electron
    • SiliconValleyGarage
« Reply #2 on: April 24, 2013, 03:04:50 pm »
Most likely that big power hybrid is shot. On the bottom board there is an industrial power converter brick. Either a vicor or an ericsson. They are hard to find...
Offline richcj10
« Reply #3 on: April 24, 2013, 03:09:07 pm »
It works.
I get the 12v and 5v out of it.
I just discovered if I run my o-scope probe across one side of U3 (Main CPU)
it will reset and turn on with a "No Com" error.