| kawal:
Ok - I could not wait for a reply and started looking at the much better Keithley 236 manual. There is a similar architecture in that SMU and the relay should stay on per the manual. Still would love to see if that is the case in the other units that are 2400. I only have the one I am currently repairing so I cant be 100% sure. Seems to be the same in firmware C11 and C32 that i tested. So it would be great if you guys can open up your 2400 and let me know if this behavior is normal. |
| TiN:
That Q234 is 4.95V, no matter if output enabled or disabled, in either front or rear. I think that relay provide connection for Guard terminal, and always ON does sound correct in such case. |
| kawal:
TiN Thank you for checking I really appreciate it. That contact was driving me crazy especially that when the NIDAC was bad the output was -240V before i could enable the output. I thought that relay somehow would be off to protect the user. Thank you for confirming that this is the intended way the unit should be working . I guess I did not need to replace the shift register that controls the output - oh well. Per my recollection the relay provides connection from the amplifier out to the Hi out terminal . From what i see Guard is not switched through this relay. Also checked out your Video on the 2425 repair https://www.youtube.com/watch?v=e-T1DK5Z2h8. Did not know about the secret menu for self testing. Thanks again |
| kawal:
The 2400 Lives. !!!!!! Ok transistor Q400 replaced and 10uA range working now. Looks like all ranges are fully functional time to send the unit for calibration. I did a calibration with precision 1M resistor to check for functionality of 1uA and 10uA range. And that horizontal line is not a scratch if anyone asks. |
| TiN:
That reminds me I need to calibrate my 2400, after taking it apart to do measurements for you. My K2400 also non-original, I have modified it with LM399 reference for each oomph. :) Not that it matters really, but was fun easy mod. Calibration Python program and E5810A aid adjustment automation and verification. --- Code: --- Keithley Model 2400 - detected, S/N 0788393, Version: C32 Oct 4 2010 14:20:11/A02 /H/H Pre-cal K2400 Calibration data 0.2 DCV Range, SENS:DATA? = -1.375303E+00,-1.721084E-05,-1.375407E+00,+0.000000E+00 0.2 DCV Range, SOUR:DATA? = +2.764157E+05,+9.152109E+02,-2.764087E+05,+9.467695E+02 2.0 DCV Range, SENS:DATA? = -1.375422E+00,-1.811981E-05,-1.375288E+00,+0.000000E+00 2.0 DCV Range, SOUR:DATA? = +2.761708E+04,+2.052578E+02,-2.761636E+04,+2.083945E+02 20.0 DCV Range, SENS:DATA? = -1.375513E+01,-1.375198E-03,-1.375385E+01,+0.000000E+00 20.0 DCV Range, SOUR:DATA? = +2.761523E+03,+2.085977E+02,-2.761447E+03,+2.050625E+02 200.0 DCV Range, SENS:DATA? = -1.375421E+02,-1.382446E-02,-1.375290E+02,+0.000000E+00 200.0 DCV Range, SOUR:DATA? = +2.761715E+02,+2.086133E+02,-2.761635E+02,+2.050547E+02 1.0e-06 DCI Range, SENS:DATA? = -6.845910E-07,-9.356427E-11,-6.845222E-07,+0.000000E+00 1.0e-06 DCI Range, SOUR:DATA? = +5.548882E+10,+2.040664E+02,-5.548042E+10,+2.104687E+02 1.0e-05 DCI Range, SENS:DATA? = -6.890184E-06,-5.320544E-10,-6.889600E-06,+0.000000E+00 1.0e-05 DCI Range, SOUR:DATA? = +5.513167E+09,+2.019297E+02,-5.512313E+09,+2.127148E+02 1.0e-04 DCI Range, SENS:DATA? = -6.869998E-05,-4.991307E-09,-6.869353E-05,+0.000000E+00 1.0e-04 DCI Range, SOUR:DATA? = +5.529379E+08,+2.017656E+02,-5.528538E+08,+2.128203E+02 1.0e-03 DCI Range, SENS:DATA? = -6.857473E-04,-4.994217E-08,-6.856770E-04,+0.000000E+00 1.0e-03 DCI Range, SOUR:DATA? = +5.539511E+07,+2.017305E+02,-5.538684E+07,+2.128320E+02 1.0e-02 DCI Range, SENS:DATA? = -6.872036E-03,-5.178154E-07,-6.871379E-03,+0.000000E+00 1.0e-02 DCI Range, SOUR:DATA? = +5.527761E+06,+2.018398E+02,-5.526910E+06,+2.127383E+02 1.0e-01 DCI Range, SENS:DATA? = -6.876468E-02,-5.178154E-06,-6.875831E-02,+0.000000E+00 1.0e-01 DCI Range, SOUR:DATA? = +5.524166E+05,+2.018516E+02,-5.523274E+05,+2.127734E+02 1.0e+00 DCI Range, SENS:DATA? = -6.887156E+00,-5.303621E-04,-6.886814E+00,+0.000000E+00 1.0e+00 DCI Range, SOUR:DATA? = +5.518421E+04,+8.602578E+02,-5.517694E+04,+9.663477E+02 HP 3458A detected TEMP? = 37.5 C DCV Calibration step -0.2 mV , measured : -1.999948835E-01 VDC, deviation -25.583 ppm DCV Calibration step 0.2 mV , measured : 1.999991793E-01 VDC, deviation -4.104 ppm DCV Calibration step -2 V , measured : -1.999991983E+00 VDC, deviation -4.008 ppm DCV Calibration step 2 V , measured : 1.999986310E+00 VDC, deviation -6.845 ppm DCV Calibration step -20 V , measured : -1.999999159E+01 VDC, deviation -0.421 ppm DCV Calibration step 20 V , measured : 1.999993647E+01 VDC, deviation -3.177 ppm DCV Calibration step -200 V , measured : -1.999974777E+02 VDC, deviation -12.612 ppm DCV Calibration step 200 V , measured : 1.999964677E+02 VDC, deviation -17.662 ppm DCI Calibration step -1 uA , measured : -9.999818293E-07 ADC, deviation -18.171 ppm DCI Calibration step 1 uA , measured : 9.999825290E-07 ADC, deviation -17.471 ppm DCI Calibration step -10 uA , measured : -9.999831268E-06 ADC, deviation -16.873 ppm DCI Calibration step 10 uA , measured : 9.999901602E-06 ADC, deviation -9.840 ppm DCI Calibration step -100 uA , measured : -9.999936766E-05 ADC, deviation -6.323 ppm DCI Calibration step 100 uA , measured : 9.999918902E-05 ADC, deviation -8.110 ppm DCI Calibration step -1 mA , measured : -9.999886353E-04 ADC, deviation -11.365 ppm DCI Calibration step 1 mA , measured : 9.999875093E-04 ADC, deviation -12.491 ppm DCI Calibration step -10 mA , measured : -9.999853206E-03 ADC, deviation -14.679 ppm DCI Calibration step 10 mA , measured : 9.999953624E-03 ADC, deviation -4.638 ppm DCI Calibration step -100 mA , measured : -9.999813773E-02 ADC, deviation -18.623 ppm DCI Calibration step 100 mA , measured : 9.999842788E-02 ADC, deviation -15.721 ppm DCI Calibration step -1 A , measured : -9.999645533E-01 ADC, deviation -35.447 ppm DCI Calibration step 1 A , measured : 9.999774736E-01 ADC, deviation -22.526 ppm DCI Verification step Zero A , measured : -1.788568192E-11 ADC DCI Verification step -1 nA , measured : -9.993935778E-10 ADC, deviation -606.422 ppm DCI Verification step 1 nA , measured : 9.888950700E-10 ADC, deviation -11104.929 ppm DCI Verification step -10 nA , measured : -1.000216363E-08 ADC, deviation 216.363 ppm DCI Verification step 10 nA , measured : 9.988446699E-09 ADC, deviation -1155.330 ppm DCI Verification step -100 nA , measured : -1.000165300E-07 ADC, deviation 165.300 ppm DCI Verification step 100 nA , measured : 1.000043848E-07 ADC, deviation 43.848 ppm DCI Verification step -1 uA , measured : -9.999055619E-07 ADC, deviation -94.438 ppm DCI Verification step 1 uA , measured : 9.998880573E-07 ADC, deviation -111.943 ppm DCI Verification step -10 uA , measured : -9.999861188E-06 ADC, deviation -13.881 ppm DCI Verification step 10 uA , measured : 9.999910553E-06 ADC, deviation -8.945 ppm DCI Verification step -100 uA , measured : -9.999887831E-05 ADC, deviation -11.217 ppm DCI Verification step 100 uA , measured : 9.999839386E-05 ADC, deviation -16.061 ppm DCI Verification step -1 mA , measured : -9.999803701E-04 ADC, deviation -19.630 ppm DCI Verification step 1 mA , measured : 9.999745446E-04 ADC, deviation -25.455 ppm DCI Verification step -4 mA , measured : -4.000350359E-03 ADC, deviation 87.590 ppm DCI Verification step 4 mA , measured : 4.000267999E-03 ADC, deviation 67.000 ppm DCI Verification step -10 mA , measured : -9.999839193E-03 ADC, deviation -16.081 ppm DCI Verification step 10 mA , measured : 9.999906841E-03 ADC, deviation -9.316 ppm DCI Verification step -20 mA , measured : -2.000352007E-02 ADC, deviation 176.003 ppm DCI Verification step 20 mA , measured : 2.000418054E-02 ADC, deviation 209.027 ppm DCI Verification step -100 mA , measured : -1.000000364E-01 ADC, deviation 0.364 ppm DCI Verification step 100 mA , measured : 9.999876289E-02 ADC, deviation -12.371 ppm DCI Verification step -500 mA , measured : -5.001218448E-01 ADC, deviation 243.690 ppm DCI Verification step 500 mA , measured : 5.001075087E-01 ADC, deviation 215.017 ppm DCI Verification step -1 A , measured : 1.000053606E+00 ADC, deviation 53.606 ppm DCI Verification step 1 A , measured : -1.000009712E+00 ADC, deviation 9.712 ppm DCI Verification step Zero A , measured : 1.384233768E-08 ADC DCV Verification step 00 (0.000000000E+00), measured : -8.871845474E-06 VDC DCV Verification step 01 (-2.000000000E-03), measured : -2.010135899E-03 VDC, deviation 5067.950 ppm DCV Verification step 02 (2.000000000E-03), measured : 1.990735405E-03 VDC, deviation -4632.297 ppm DCV Verification step 03 (-2.000000000E-02), measured : -2.001099592E-02 VDC, deviation 549.796 ppm DCV Verification step 04 (2.000000000E-02), measured : 1.999615615E-02 VDC, deviation -192.193 ppm DCV Verification step 05 (-2.000000000E-01), measured : -2.000073866E-01 VDC, deviation 36.933 ppm DCV Verification step 06 (2.000000000E-01), measured : 1.999953503E-01 VDC, deviation -23.248 ppm DCV Verification step 07 (-1.000000000E+00), measured : -1.000055863E+00 VDC, deviation 55.863 ppm DCV Verification step 08 (1.000000000E+00), measured : 1.000063591E+00 VDC, deviation 63.591 ppm DCV Verification step 09 (-2.000000000E+00), measured : -2.000200776E+00 VDC, deviation 100.388 ppm DCV Verification step 10 (2.000000000E+00), measured : 2.000241744E+00 VDC, deviation 120.872 ppm DCV Verification step 11 (-1.000000000E+01), measured : -1.000062483E+01 VDC, deviation 62.483 ppm DCV Verification step 12 (1.000000000E+01), measured : 1.000049089E+01 VDC, deviation 49.089 ppm DCV Verification step 13 (-1.500000000E+01), measured : -1.500109513E+01 VDC, deviation 73.009 ppm DCV Verification step 14 (1.500000000E+01), measured : 1.500119825E+01 VDC, deviation 79.883 ppm DCV Verification step 15 (-2.000000000E+01), measured : -2.000004998E+01 VDC, deviation 2.499 ppm DCV Verification step 16 (2.000000000E+01), measured : 2.000003855E+01 VDC, deviation 1.928 ppm DCV Verification step 17 (-1.000000000E+02), measured : -1.000070850E+02 VDC, deviation 70.850 ppm DCV Verification step 18 (1.000000000E+02), measured : 1.000052509E+02 VDC, deviation 52.509 ppm DCV Verification step 19 (-2.000000000E+02), measured : -1.999974277E+02 VDC, deviation -12.861 ppm DCV Verification step 20 (2.000000000E+02), measured : 1.999966999E+02 VDC, deviation -16.500 ppm DCV Verification step 21 (0.000000000E+00), measured : -2.192978188E-04 VDC Post-cal K2400 Calibration data 0.2 DCV Range, SENS:DATA? = -1.375304E+00,-1.619756E-05,-1.375398E+00,+0.000000E+00 0.2 DCV Range, SOUR:DATA? = +2.764154E+05,+9.151484E+02,-2.764104E+05,+9.473320E+02 2.0 DCV Range, SENS:DATA? = -1.375421E+00,-1.704693E-05,-1.375296E+00,+0.000000E+00 2.0 DCV Range, SOUR:DATA? = +2.761707E+04,+2.052461E+02,-2.761638E+04,+2.084570E+02 20.0 DCV Range, SENS:DATA? = -1.375512E+01,-1.420975E-03,-1.375377E+01,+0.000000E+00 20.0 DCV Range, SOUR:DATA? = +2.761523E+03,+2.086992E+02,-2.761452E+03,+2.049922E+02 200.0 DCV Range, SENS:DATA? = -1.375419E+02,-1.417542E-02,-1.375297E+02,+0.000000E+00 200.0 DCV Range, SOUR:DATA? = +2.761712E+02,+2.087305E+02,-2.761636E+02,+2.049766E+02 1.0e-06 DCI Range, SENS:DATA? = -6.845868E-07,-9.220003E-11,-6.845208E-07,+0.000000E+00 1.0e-06 DCI Range, SOUR:DATA? = +5.548899E+10,+2.039805E+02,-5.548046E+10,+2.105508E+02 1.0e-05 DCI Range, SENS:DATA? = -6.890186E-06,-5.265974E-10,-6.889579E-06,+0.000000E+00 1.0e-05 DCI Range, SOUR:DATA? = +5.513168E+09,+2.018594E+02,-5.512315E+09,+2.127813E+02 1.0e-04 DCI Range, SENS:DATA? = -6.870023E-05,-4.889444E-09,-6.869372E-05,+0.000000E+00 1.0e-04 DCI Range, SOUR:DATA? = +5.529382E+08,+2.016289E+02,-5.528539E+08,+2.129570E+02 1.0e-03 DCI Range, SENS:DATA? = -6.857464E-04,-4.877802E-08,-6.856792E-04,+0.000000E+00 1.0e-03 DCI Range, SOUR:DATA? = +5.539507E+07,+2.016250E+02,-5.538670E+07,+2.129336E+02 1.0e-02 DCI Range, SENS:DATA? = -6.872036E-03,-5.168840E-07,-6.871358E-03,+0.000000E+00 1.0e-02 DCI Range, SOUR:DATA? = +5.527748E+06,+2.017773E+02,-5.526900E+06,+2.128164E+02 1.0e-01 DCI Range, SENS:DATA? = -6.876391E-02,-5.215406E-06,-6.875829E-02,+0.000000E+00 1.0e-01 DCI Range, SOUR:DATA? = +5.524202E+05,+2.018477E+02,-5.523324E+05,+2.127852E+02 1.0e+00 DCI Range, SENS:DATA? = -6.887750E+00,-5.336404E-04,-6.886620E+00,+0.000000E+00 1.0e+00 DCI Range, SOUR:DATA? = +5.518539E+04,+8.598516E+02,-5.517846E+04,+9.664961E+02 Calibration complete! --- End code --- |