Hello,
It's my first post so hello to everybody :)
I recived Keithley daq6510 and i've error with code: 5526 it says: Card Calibration data lost, temperature readings may be compromised. I attached screenshoot with whole description of it. Anybody met with it? I already tried to calibrate "cold temperature" via SCPI but it doesn't change anything. :/ Maybe somebody have got that module and can send me buffer of 24C164 EEPROM? There is one topic with eeproms of 7703/7705 modules and if I put the good one, then DAQ reconize it as 7705 without this error, so I think the fault is connected with it.
Best Regards :)
-
Welcome to the forum.
Don't you have full warranty on the multiplexer module?
How old are the DAQ6510 and the 7700?
-
Unfortunately no :/ DAQ6510 is new, but 7700 is used and old :<
-
I have a 7700, but not in a position where I can attempt to pull the ROM data off of it. Sadly.....
I think there are a few others that have had troubles with these - hopefully, someone already has an image available.
-
Hi,
a while back I had created EEPROM backups of my multiplexer cards just for fun :)
Attached is the bin file of one of my 7700 multiplexer cards.
I got a second one but from this one I don't have created a dump yet.
-
Hello,
thank you for your contribution! It was very helpful to make my non-recognized card recognizable :-+. Then I did a backup of my 7702 eeprom for the case this might fail in the future.