No tool can do it all. In this case, DMM6550 clearly fails measuring R in inductive loads. My old Fluke 8842A doesn’t have any issue with that.
However, it’s not as bad as it seems. It looks to me like the issue is mostly related to autoranging, rather than to the measurement itself. Once set in manual range, the measurement seems quite accurate (according to my quick tests).
I measured the primary coil of a mains transformer, checked as follows:
L = 2.2 H @ 100 Hz (DE-5000)
DCR = 63.19 ohms (Fluke 87V)
DCR = 63.10 ohms (Brymen 789)
R = 63.104 ohms (old Fluke 8842A, autorange, with no issues)
DMM6550 measured as follows:
R = 00.00006 M (autorange) (actually ranging from 00.00004 M to 00.00007 M)
R = 00.00006 M (10M manual range)
R = 0.000067 M (1M manual range)
R= 00.06318 K (100K manual range)
R = 0.063183 K (10K manual range)
R = 0.063173 K (1K manual range)
Overflow (100 manual range and under)
The most interesting things happened when I checked the signal with an oscilloscope.
In autorange, DMM6550 shows some huge (40Vpp) spikes, which miraculously disappear in manual range.
Autorange:



Manual range:

In manual range, the magnified signal looks very similar with the on from Fluke 8842:
Edit: LED lamp artifact (Thanks, Kleinstein).
Might be wishful thinking, but what is the chance this is just another fixable software bug?