Products > Test Equipment
Keysight 3000T scope 1 GHz(or more) hardware upgrade mod
electrolust:
--- Quote from: TheSteve on June 30, 2017, 04:03:13 am ---#17 - The 50 ohm path is much improved with the updated front end and use of the Teledyne 20 dB attenuator relays
--- End quote ---
Would this part of the mod be worth doing just on its own, for a 200MHz scope, without doing the rest of the mod? Or at 200MHz is it not gaining you so much.
Amazing:
Great job! that really took guts.
--- Quote from: Keysight_DanielBogdanoff on June 30, 2017, 03:45:46 pm ---scope. And, as is made clear by the mod, the whole backend of the scope is capable to 1 GHz (or higher >:D)
--- End quote ---
The first version of the 3000T manual mentions a 1.5GHz model that was never sold...
TheSteve:
--- Quote from: Dubbie on June 30, 2017, 11:22:36 am ---I really am keen to try this but I am not sure I have the courage. Can you do it to a single channel? Or is is all or none?
--- End quote ---
This is an all or nothing mod and if you aren't sure please don't risk your scope.
--- Quote from: Keysight_DanielBogdanoff on June 30, 2017, 03:45:46 pm ---SNIP
And, as is made clear by the mod, the whole backend of the scope is capable to 1 GHz (or higher >:D)
SNIP
--- End quote ---
Is that a challenge?
I know the front end ASIC is capable of at least 1.5 GHz as it is used in the 4000x series. My dream was a 1.5 GHz upgrade but without seeing a 4000 series 1.5 GHz front end it just isn't going to happen. It was hard enough to determine the required values for 1 GHz.
There is the jumper option on the board for 1.5 GHz but as far as I am aware it will be rejected as it isn't a 4000 series board. I have no desire to get into firmware hacking.
I don't need to sleep, I'll be pondering it.
--- Quote from: electrolust on June 30, 2017, 06:06:51 pm ---Would this part of the mod be worth doing just on its own, for a 200MHz scope, without doing the rest of the mod? Or at 200MHz is it not gaining you so much.
--- End quote ---
There is nothing to be gained by modifying the 200 MHz front end for a better 50 ohm match. It also wouldn't work properly as the relay/attenuator controls change when you go from 100-500 to 1 GHz hardware.
--- Quote from: electrolust on June 30, 2017, 05:58:15 pm ---Can you characterize the response curve? My memory might be bad, but as I recall from the literature, the sub-1GHz models have gaussian response and the 1GHz (as shipped from Keysight) has maximally flat response. Apologies if the posted traces show that fact, I'm a bit excited and haven't read through in detail!
--- End quote ---
I do believe the 1 GHz model has the flat response. If you look at the anti alias/low pass filter on the output of the front end ASIC you'll see the 1 GHz model has an added series resistor/capacitor. In simulation this prevents attenuation above 500 MHz and changes the curve around 1 GHz.
I will plot the curve when I get a chance.
KE5FX:
WastelandTek:
wowwwww...