Hi Samofab,
Thanks for your reply. Although my display shows a raster, and the backlight and contrast controls work properly, the unit is not responsive in that there is no menu, text or cursors. The buttons on the display have no effect, so I can't run further self-tests via the display. I've just reassembled the display unit after inspecting the buttons for physical faults, but all seems fine. A check of the voltages on the display cable connector revealed plus and minus 15.0 volts, zero on the ground pins, and various DC voltages all less than 5 v. No sign of a 4 MHz sine wave. I don't think my problem is within the display unit itself.
Older Tek 'scopes could often be "brought to their knees" by a single faulty tantalum bypass capacitor somewhere across a DC supply. But there aren't many tantalums in the RTD 720, and they all tested OK. So it seems that I must start a serious hunt for a service manual. Do you remember who and how you made contact with Tek in your search for a manual? Perhaps an old-fashioned snail mail letter to their chief engineer might bear more fruit, as I simply don't believe that they no longer have at least one copy of a service manual, not only for the RTD720A, but for every single instrument they've ever produced.
Realising that you won't be too interested in these instruments now that you've sold yours, are there any more clues you might have, or any other contact people I should try to get in touch with? Many thanks.
Final question - the Operator Manual mentions an internal battery which is tested in the power-up diagnostic. Did you happen to discover where this particular battery is located? I didn't notice any battery when I dismantled the various circuit boards of the main unit, but I wasn't specifically searching for one at that stage.