Products > Test Equipment
keysight dmm 34461a data logging problem
HKJ:
--- Quote from: robint on October 15, 2023, 12:18:46 am ---
;; 192.168.1.229 Device Keysight Technologies,34461A, do not match: Agilent Technologies,34461A,MY53200769,A.03.03-02.40-03.03-00.52-01-01
So I dont know what to do next? I suspect some devious Keysight stuff? Is there some way to test my ehternet link ? I had a lots of trouble with Keysight 10 years ago trying to make it work with Benchvue and usb - wasted months with BS from their head office. USB was a new thing for them, badly implemented.
--- End quote ---
The issue is that your meter is a Agilent, not a Keysigt meter.
It is easy to fix:
1) Download the attached file
2) Replace the similar named file in the TestController/Devices directory with it.
3) Delete the Keysigt entry in your device list
4) Add the Agilent entry instead (It is defined in the new file).
Note: I have only added the 34461A, if anybody wants the 65A or 70A I can also add them
coromonadalix:
would be nice to put this in a text file instead of a lenghty one here ???
robint:
:)
Just trying to be precise and polite, because we are dealing with two files with the same name but different content - bad news :rant:
so the post was intended for HJK
HKJ:
--- Quote from: robint on October 15, 2023, 02:46:15 pm ---Done that replacement, still get same error?
;; 192.168.1.229 Device Keysight Technologies,34461A, do not match: Agilent Technologies,34461A,MY53200769,A.03.03-02.40-03.03-00.52-01-01
;; Use "Load devices" page for selecting devices to load. Most devices will only load if they are connected and ready.
--- End quote ---
You did not replace the Keysight device on the "Load devices" page with a Agilent.
The error messages means that the #idString line do not match what the device returns and because it still says Keysigt you have not loaded the Agilent definition.
--- Quote from: robint on October 15, 2023, 02:46:15 pm ---just so we are on the same page
here's what I put in to the devices folder
--- End quote ---
Please, dont post the full file, it is a lot of noise in the thread.
robint:
Ok 41 sheets of code is gruesome :horse: but you gave me a dilemma with a filename with two different contents - but I did the replacement in the directory
I thought you said that when TC starts up it automatically detects what instrument it is connected to?
So It now comes up with
Agilent 3661A
Socket
Address sn:MY53200769
Baud -
Enable ok
Notes
Does this mean its looking for an Ethernet cable and not finding USB?
I did find the Agilent in the "search socket " window " and selected that and it now comes up by default but still not displaying any values?
Your system doesnt give info as to whether a successful connection has been made eg IDN?
I finally got the original Benchvue loaded and it seems to work and log - I will post a separate topic on this - its been a very very fraught journey and its really hidden away and FREE as it was originally